Mistery flowmarks on C41 negatives from dip and dunk

Recent Classifieds

Forum statistics

Threads
201,983
Messages
2,833,290
Members
101,048
Latest member
simenswang
Recent bookmarks
0

owk.edo

Member
Joined
Jan 26, 2026
Messages
9
Location
London
Format
Medium Format
Hello! Posting here for the first time and in search of some help on an issue I've been having in my C41 process.

I've been running a Tecnloab dip and dunk for 4 years now, it's been serviced by a professional when first set up at my lab and I've hardly made any changes to the setup, times or workflow.
I use Champion chemistry and monitor sensitometry daily, with pretty consistent results or at least nothing that ever granted for major adjustments.

Since a couple of months I've started noticing these flowmarks on the negatives (attaching scan below from unexposed MF Portra 400 I used as test), they are clearly visible on less dense parts of the negs and sometimes even to the naked eye on the base when I look at them on the light table.

I did some troubleshooting with the knowledge I have, but feel like I need the help of someone more experienced in the use of these machines.

This is what I did
- skip the stabilizer bath: I was afraid of the stabilizer being overly concentrated/not replenished enough so I skipped it on a couple of racks and the marks were still there.
- skip the dryer: I pulled a few rolls out of the machine right before the dryer and dried them at room temp in the cabinet I use for B/W. Same result.

I am now afraid that this has to do with agitation in either the dev or the bleach. My sensitometry tells me the agitation is correct but I feel like I might be missing something.

Any help/input will be greatly appreciated. Thank you!

E
 

Attachments

  • Schermata 2026-01-26 alle 11.07.53.png
    Schermata 2026-01-26 alle 11.07.53.png
    1.7 MB · Views: 38

koraks

Moderator
Moderator
Joined
Nov 29, 2018
Messages
27,162
Location
Europe
Format
Multi Format
I'm linking also this thread which involves a D&D system and problems with final wash: https://www.photrio.com/forum/threads/circular-and-hazy-stains-in-dip-and-dunk-c-41.217924 and perhaps @vicb is willing to have a look at your problem as well on the basis of his experience with D&D processing.


I am now afraid that this has to do with agitation in either the dev or the bleach.
It's easy enough to figure out there the problem originates. Take two samples of affected film. Re-fix one sample. Re-bleach and re-fix another sample. If both show the same problem, it existed before bleach and thus must be caused during development. If both look fine, the problem is in the fix. If the re-bleached + re-fixed sample is OK and the re-fixed one isn't, the problem is in the bleach.

The first thing that comes to mind is indeed something associated with agitation and I would expect that the problem is in development, so my money goes to re-bleaching and re-fixing not showing any change. If it's not an agitation problem in the development tank, then I would guess that it's a replenishment and/or carry-over problem in either of the subsequent ones, depending on how the strip test goes.

Btw, your attached sample also shows quasi-circular drying spots but I assumed above that we should ignore those for now.
 
OP
OP
owk.edo

owk.edo

Member
Joined
Jan 26, 2026
Messages
9
Location
London
Format
Medium Format
I'm linking also this thread which involves a D&D system and problems with final wash: https://www.photrio.com/forum/threads/circular-and-hazy-stains-in-dip-and-dunk-c-41.217924 and perhaps @vicb is willing to have a look at your problem as well on the basis of his experience with D&D processing.



It's easy enough to figure out there the problem originates. Take two samples of affected film. Re-fix one sample. Re-bleach and re-fix another sample. If both show the same problem, it existed before bleach and thus must be caused during development. If both look fine, the problem is in the fix. If the re-bleached + re-fixed sample is OK and the re-fixed one isn't, the problem is in the bleach.

The first thing that comes to mind is indeed something associated with agitation and I would expect that the problem is in development, so my money goes to re-bleaching and re-fixing not showing any change. If it's not an agitation problem in the development tank, then I would guess that it's a replenishment and/or carry-over problem in either of the subsequent ones, depending on how the strip test goes.

Btw, your attached sample also shows quasi-circular drying spots but I assumed above that we should ignore those for now.

Hello Koraks, thank you very much for the reply. I did stumble upon the thread you linked when I was googling for issues like mine and, although it offers some useful insights on the drying process, I believe my issue is something entirely different.

The drying spots occurred because I dried these tests way too fast as I was eager to inspect them. No issues with the dryer so far.

I'll be running the re-bleach and re-fix tests, thank you for the suggestion.

I would guess that it's a replenishment and/or carry-over problem in either of the subsequent ones, depending on how the strip test goes

It did cross my mind that I could have a carry-over issue between the stabilizer and the developer: the racks and clips I'm using have been custom made for me after I back engineered the only original rack I have (the person I bought the machine from threw away all the racks and clips he had). When I pull them out of the dryer they often have some wet residue and they usually go straight back through the machine as I don't have loads.
 

koraks

Moderator
Moderator
Joined
Nov 29, 2018
Messages
27,162
Location
Europe
Format
Multi Format
Ah I see. It's definitely possible that cross-contamination plays a role here. You esp. don't want any bleach or fix to end up in the developer. Via the stab bath I could see either (esp. fix) carry over into the developer. Stabilizer itself in the developer is likely not great either. So this is definitely something to eliminate if you suspect it plays a role. To verify you'd have to start over with a fresh developer tank as it's generally not possible to regenerate/repair a contaminated developer bath.
 
OP
OP
owk.edo

owk.edo

Member
Joined
Jan 26, 2026
Messages
9
Location
London
Format
Medium Format
I see your point and from the information I've given it makes sense to come to that conclusion but I'd like to point out that:

1. wouldn't that amount of cross contamination be well visible in my sensitometry? My control strips are within range and - apart from the flowmarks - the negatives are clean and print well
2. I've been running this setup for 4ish years now without any major changes in the workflow and this issue just started arising in the last few months. I must be missing /overlooking something but I can't think of anything I could have done different that started this

I appreciate your input, I'll be running the tests you recommended and if the result is unchanged I'll probably bite the bullet and make a fresh tank of developer.
 

koraks

Moderator
Moderator
Joined
Nov 29, 2018
Messages
27,162
Location
Europe
Format
Multi Format
Yeah, same here, I see your points as well! Let's see what the strip tests say; maybe they will clarify things a bit.

Btw - if it's carryover from stab into dev I can sort of see how it wouldn't shop up on densitometry if the effect remains limited to uneven wetting at a micro scale. But given the snippet you showed above with all the strips I would frankly expect you see this variation in the densitometry as well if you take several measurements of the same patch.
 

koraks

Moderator
Moderator
Joined
Nov 29, 2018
Messages
27,162
Location
Europe
Format
Multi Format
Couldn't this be caused by insufficient mixing (stirring when mixing) of the chemistry?
No way. If you mix solutions (which are what the concentrates used in C41 already are), they diffuse. Even if you tried, you couldn't keep them neatly separated like this when mixed together. This is a fundamental law of thermodynamics.
Even if we ignored that (i.e. we now are talking magic and no longer reality/science), the stuff would still mix because it's a D&D line so there's going to be agitation, typically using nitrogen burst. That will mix the contents of the bath in no time.
 

John Salim

Member
Allowing Ads
Joined
Apr 15, 2014
Messages
272
Location
Southend Essex
Format
Multi Format
No way. If you mix solutions (which are what the concentrates used in C41 already are), they diffuse. Even if you tried, you couldn't keep them neatly separated like this when mixed together. This is a fundamental law of thermodynamics.
Even if we ignored that (i.e. we now are talking magic and no longer reality/science), the stuff would still mix because it's a D&D line so there's going to be agitation, typically using nitrogen burst. That will mix the contents of the bath in no time.

... plus the machine will be running circulation pumps, so constantly being mixed.

John S 😊
 

Kino

Subscriber
Allowing Ads
Joined
Jan 20, 2006
Messages
7,975
Location
Orange, Virginia
Format
Multi Format
When you "dump tank" on the developer, I would assume you do clean/flush the tanks and change the recirculation filters, right?

Not trying to be insulting here, but the contamination probably won't go away unless you do this.
 
OP
OP
owk.edo

owk.edo

Member
Joined
Jan 26, 2026
Messages
9
Location
London
Format
Medium Format
When you "dump tank" on the developer, I would assume you do clean/flush the tanks and change the recirculation filters, right?

Not trying to be insulting here, but the contamination probably won't go away unless you do this.

Yes of course, if it comes to that I'll make sure everything is replaced and cleaned before filling up again with fresh solution
 
OP
OP
owk.edo

owk.edo

Member
Joined
Jan 26, 2026
Messages
9
Location
London
Format
Medium Format
Yeah, same here, I see your points as well! Let's see what the strip tests say; maybe they will clarify things a bit.

Just ran the two tests, I let both strips sit in the tanks for a while and then ran the machine to make sure bleaching and fixing processes achieved completion.
Unfortunately the marks are still there.

I am now wondering if there is anything else I could do before dumping the dev. Especially in order to understand what caused the issue and avoid spoiling a brand new 60L tank of dev by making the same mistake.. I might play around with agitation times and strength to see if that has any effect, even though those were set by the engineer that set the machine up 4 years ago (I did change the dev agitation over time but bleach and fix were untouched).

Have any of the gas burst plenums got slightly displaced or partially blocked?

Thank you for the suggestion, I inspected them this morning and they seem unobsctructed and sitting properly at the bottom of the tanks.
 

koraks

Moderator
Moderator
Joined
Nov 29, 2018
Messages
27,162
Location
Europe
Format
Multi Format
I don't believe that this is related to agitation to be honest.
Whatever is happening must either happen during development, or it's something that happens to the film that bleaching and fixing doesn't resolve. I can't think of anything in the latter category that fits the example you've shown, and that's no inherent to the developer bath.

That leaves the important question whether replacing the chemistry in the developer tank will actually help. Lacking a thorough analysis of what the probable cause is, I'd be at least as hesitant as you are. And it does remain odd that densitometry doesn't show any problems.

Can we have another look at the phenomenon, also on uninverted photos of the film photographed against a backlight? I'd like to have a bit of a feeling for how large the features are and what their approximate density variations are. It would also help to know if the effect is identical across the entire film surface or whether there's variation.
 
OP
OP
owk.edo

owk.edo

Member
Joined
Jan 26, 2026
Messages
9
Location
London
Format
Medium Format
I see your point, and I'd definitely prefer to do some more investigation before replacing the chemistry tbh.

Here is a link to the inverted scans of the full length of the original example I sent (2 rolls of 120 Portra 400) and pictures against a backlight of the same two negatives, taken halfway down the length and at the end - https://we.tl/t-O90VAXI1sl (please disregard the drying marks for now)

Some more elements I'd like to mention for your consideration:
- processing volume has been low during the last few months (roughly 100/120 rolls per week)
- I suspect a leak in the nitrogen line as sometimes the level on the tank decreases overnight (but no audible leak detected)

Thank you very much for taking the time!
 

koraks

Moderator
Moderator
Joined
Nov 29, 2018
Messages
27,162
Location
Europe
Format
Multi Format
Thanks for posting the additional illustrations. For anyone interested in the same things I am, I'll post the (to me) interesting bits here if you don't mind:

1769527230986.png

1769527253153.png


This is part of the first strip in the top image with dramatically increased contrast (but no inversion):
1769527399762-png.416654


So logitudinal stripes of varying density with, no particular color.
First strip in the second photo (with the curled end) seems to be much less severely affected. Is there a pattern to the severity along the length of the roll?
Does the problem also occur to the same degree on 135? If you also have 135 that's affected, is there anything odd going on around the sprocket holes?

I think a contamination problem of the developer is still a likely scenario.
 

Attachments

  • 1769527399762.png
    1769527399762.png
    854.1 KB · Views: 34
OP
OP
owk.edo

owk.edo

Member
Joined
Jan 26, 2026
Messages
9
Location
London
Format
Medium Format
Thank you for looking into this.

First strip in the second photo (with the curled end) seems to be much less severely affected. Is there a pattern to the severity along the length of the roll?

Apologies, I think that might be due to the top strip being curled and showing less of the effect in this particular picture I took. If I lay it flat next to the other one they appear to be affected in the same way.
As far as I can tell there is no pattern, the marks appear to follow the way the chemistry drips down the roll while it's being lifted maybe? They're hung with the end of the roll at the bottom.

Does the problem also occur to the same degree on 135? If you also have 135 that's affected, is there anything odd going on around the sprocket holes?

Yes it does, I don't have an example but can probably find one. The effect is the same and in the same scale, there is nothing odd around the sprocket holes as far as I can tell.
 

John Salim

Member
Allowing Ads
Joined
Apr 15, 2014
Messages
272
Location
Southend Essex
Format
Multi Format
Hello! Posting here for the first time and in search of some help on an issue I've been having in my C41 process.

I've been running a Tecnloab dip and dunk for 4 years now, it's been serviced by a professional when first set up at my lab and I've hardly made any changes to the setup, times or workflow.
I use Champion chemistry and monitor sensitometry daily, with pretty consistent results or at least nothing that ever granted for major adjustments.

Since a couple of months I've started noticing these flowmarks on the negatives (attaching scan below from unexposed MF Portra 400 I used as test), they are clearly visible on less dense parts of the negs and sometimes even to the naked eye on the base when I look at them on the light table.

I did some troubleshooting with the knowledge I have, but feel like I need the help of someone more experienced in the use of these machines.

This is what I did
- skip the stabilizer bath: I was afraid of the stabilizer being overly concentrated/not replenished enough so I skipped it on a couple of racks and the marks were still there.
- skip the dryer: I pulled a few rolls out of the machine right before the dryer and dried them at room temp in the cabinet I use for B/W. Same result.

I am now afraid that this has to do with agitation in either the dev or the bleach. My sensitometry tells me the agitation is correct but I feel like I might be missing something.

Any help/input will be greatly appreciated. Thank you!

E

What's your nitrogen timing cycle ? ( burst time / delay time )
Does the machine burst whilst transporting ?
Is there a delay in the first gas burst as the rack lands in the dev, or does it burst as soon as it lands ?

John S
 
OP
OP
owk.edo

owk.edo

Member
Joined
Jan 26, 2026
Messages
9
Location
London
Format
Medium Format
What's your nitrogen timing cycle ? ( burst time / delay time )
Does the machine burst whilst transporting ?
Is there a delay in the first gas burst as the rack lands in the dev, or does it burst as soon as it lands ?
Hi John, thank you for the message.

Cycle is 3s ON, 7s OFF. There is also a 1s pre-agitation setting that happens just before the rack lands.
The machine does not burst while transporting but I guess the pre-agitation happens at the very last moment of the lift movement if that's what you mean.

E
 