There was a toolkit for the 5400 calibration process : colors, resolution, etc. It's referenced in the repair manual but probably long gone.
Did you do it on the 5400?
I guess it's not that different from other scanner procedure. There is a tutorial by Gleb Shtengel for the coolscan 8000/9000. You "just" need to move the board by very small increments so that the image is centered and not skewed. Feasible I guess.
Just signed up to partake in this post cause my 5400 has just started doing this exact same zigzag pattern. How interesting… I have pulled the thing apart twice previously to take care of lines that were caused by lubrication issues but this is new to me. I was going to buy one of the ones on eBay with the lens removed to do a swap but it’s still quite expensive due to shipping for UK to AU so if I can fix the problem by replacing capacitors that would be good.I've been using a Minolta 5400 scanner for a few years now. Despite having both top of the line Coolscans at home (the 5000 and 9000), I still believe the Minolta 5400 Mk1 is my best scanner when it comes to extracting as much information as possible from a 35mm negative. I have scanned numerous photos with all these scanners and while the Coolscan tend to have better colors right out of the box, the Minolta 5400 diffuse lighting really give a more pleasing and forgiving image, especially when you have some imperfections on your negatives. Also, its resolution and grain rendition is unmatched.
It is slow, it is a brick. But I really believe it was the pinnacle in terms of image quality when it comes to advanced user scanners for 35mm.
Anyway, after these praises for a 20 years old scanner nobody asked for, here is my problem.
I noticed in dark areas the presence of zigzag artefacts. No matter what value of multisampling I choose, they're still here. I know that Coolscans tend to suffer from the same illness. I have an old Coolscan 4000 that suffers from it. But this issue has been well documented and investigated by the community, and the culprit identified. It's a simple 30 cents capacitor involved in the Analog Digital conversion that needs to be replaced.
Unfortunately, the 5400 doesn't attract the same interest from users nowadays. Sadly, their only appeal seems to be their lens, hence the sheer number of units sold without it on Ebay. Anyway, I was wondering if anyone had any idea of what capacitor or IC chip could be falty ?
Here are two examples. One in the shadows of a slide. Another in the densest highlights of a negative. Click to enlarge :
View attachment 336889 View attachment 336890
I plan to have all the chemical capacitors replaced at some point but I wondered if anyone had already encoutered this issue and had some clues. Thanks.
But in difficult conditions, its shortcomings seem more apparent than before.
Does every software (Vuescan, Minolta DiMage Scan, Silverfast) produce same pattern?
Vuescan but it weirdley duplicates adjacent pixels and create weird 4 or 5 pixels wide streaks.
I am very grateful for @koraks attempt at changing the capacitor on the A/D board.
Yes, you demonstrated that quite convincingly with a scan of a few years ago and a recent one. Although I think these scans were made on different machines of the same type, right? Either way, I also found the difference remarkable. I have some hopes for improvement if you manage to replace the light source.
I think @Archiloque scans with the Minolta software and I tried it with Vuescan, and we get the same pattern, so it seems hardware-related. My hypothesis is still that the noise is caused by either noise injected from the stepper motor into the power supply rails, and/or EMI emanating from the stepper picked up by the CCD and accompanying circuitry. It will show up mostly when dealing with very weak signals, resulting in a poor s/n ratio.
Yes, I noticed these too, and we corresponded about them.
You're very welcome; I enjoyed the excellent exchange on the technical troubleshooting part and your responsiveness in this! I'm sorry I couldn't realize a satisfactory solution at this point.
Is it happens with 2x or 4x multisample setting? From my experience it was a visibble improvement with 2x multisample in my scanner when i used it. So i always scanned with 2x on 5400II.
I also can confirm that at pixel level Vuescan software gives striped digital texture pattern compare to original Minolta scanner software. At least it was in version somewhere in 2013 when i tested it. It is not related to exposure. It is just some sort of less accurate "demosaic" variation.
So the left side (brighter) is the new scan? It actually shows less noise (though I would atribute that to it being slightly out of focus).
I made some scans myself yesterday and can observe similar pattern. It's much more prominent with grain dissolver enabled and depends on the hardware exposure, the lower the exposure the higher the noise in shadows, which is perfectly understandable. If you use software that doesn't give you (or you don't use it) control over hardware exposure then you will definitely get slightly different results even from the same slide.
The left side has less shadow detail, but the overall scan is slightly darker, or perhaps less contrasty.