I might need a new pH probe. A few questions:
1. What brand/model would you recommend for a price of around US$60?
2. Should I pay extra for a high-salinity/high-alkalinity probe?
3. Is it normal for a probe to need a few minutes to acclimate to a high sulfite solution?
My Hanna pHEP-5 probably needs a new probe. It has always needed about 4 minutes to acclimate to a sulfite solution. Once acclimated, it responded quickly to changes in pH. But this evening, the pH started dropping as I was watching it. Adding alkali failed to raise pH. Due to Hanna's poor packaging, this probe was dry when I got it a month ago, so I expected it to be a short-liver.
What you described is one of the typical symptoms of clogged or chemically deteriorated reference electrode junction.
You should get a double junction and refillable probe. Replace the fill solution every time you use the probe. It is usually best to replace the fill solution at the end of shift.
High alkalinity electrodes use slightly different glass membrane, which is important at pH 13 and above. They are not really needed for photographic applications. The place that gets shot in photographic applications is the reference junction.
Usually, high salt solutions are faster to settle/stabilize reading. If your reading doesnt stabilize in a few minutes at 25C, the reference electrode is the first place to check.
There are two common causes for unstable reading, when the electrode is perfectly normal. One is low temperature. Measurement at 20C is not very good idea for this reason. This 5C difference is big for the glass membrane. Another cause is when you clean the electrode. If you use brush or other mechanical means to clean the glass membrane, forget it. Put the electrode in 3M KCl solution and go for lunch. You wont be making any meaningful reading for about 2 hours. You can't even calibrate the system. This is yet another reason why it is best to clean the electrode at the end of each shift.
Another possible place to go wrong is moisture or conductance leak. The input impedance of the pH meter amplifier is REALLY high. Slight moisture can cause significant leakage that affects the sensitivity "slope" of the electrode. This is common with cheap pH "tester" type devices. I remember throwing away a few of those testers for this. (But all of them used gel filled reference junction, that gets killed by the test solution in a few weeks anyway.)
The least expensive electrode I know suitable for photographic work is Hanna HI-1333. It's cheap because Hanna sells it as "student lab grade" or "fish tank" or "field monitoring" use and not really rated for analytical lab use. At least that's what Hanna tech support told me years ago. That's good enough for routine darkroom use. When I had a deep tank line for film, and slot processor set up for prints, HI-1333 was used in the processing station to monitor the health of those solutions, but it never failed on me unexpectedly. I think I replaced the electrode only once in a 10 year period (but I did take very good care of it. most people kill the electrode in less than 2 years.) and the second electrode is still alive (I just took it out of the storage).