I had pretty much this exact problem. You are correct that there is a magnetic sensor (hall sensor) that detects the position of the air distributor system, though the sensor itself is very unlikely to fail as it is a robust part and is encased on silicone. The most common cause for this is that the backup battery on one of the circuit boards has leaked and damaged traces that connect the component that handles the hall sensor logic. It's an easy fix if you're comfortable with electronics, but probably pretty intimidating if you're not. I believe my fix is detailed earlier in this same thread (on an ATL-1 rather than an ATL-2, the machines have identical electronics).
Really appreciated your quickly answer, yes I'm ok for simple electronic components. I will try to change the backup battery first. thanks
.