I finally acquired a color densitometer and scratching my head looking at my results. I'm using freshly mixed Flexicolor chemistry. The reference strip is above, my readings are below:
So it looks like my B values are pretty close, but R and G are too low. Fuji has a document called C-41 technical bulletin, their version of Kodak's Z-131. When I go through the troubleshooting/correcting actions table on page 28, I do not see a use case like mine.
Low/high temperature or low/high developer activity lead to the Blue value deviating first, which is the opposite of what I'm seeing.
Any ideas what's going on here?
So it looks like my B values are pretty close, but R and G are too low. Fuji has a document called C-41 technical bulletin, their version of Kodak's Z-131. When I go through the troubleshooting/correcting actions table on page 28, I do not see a use case like mine.
Low/high temperature or low/high developer activity lead to the Blue value deviating first, which is the opposite of what I'm seeing.
Any ideas what's going on here?
There's the $300 certificate + traceable master with 0.05F accuracy, the Paterson spirit thermometer, built-in sous vide thermometer for the tempering bath, and another traceable high-accuracy digital thermo-pen with 0.1F accuracy for extra convenience. As I said, my process is accurate within 0.1F inside the tank, I had a lot of spare time during COVID. Besides, I tried raising and lowering temperature on purpose, and densities go up/down together maintaining the same blue-dominates-green spread.