If you have a calibrated step wedge from Stouffer, you can compare each reading with your Xrite 810 to the calibrated step nearest.
Sorry for the late response. Had c41 chemistry problems that left no time for densitometer.
To tell the true till I got X-rite 361T i was using calibrated Stouffer 21 to calibrate my XRite-810. I was using 3.05 field for all 3 channels buy actually only G was measured in real work (BW and C41).
With Xrite-361 I have got original xrite calibration target and used it for the calibration.
If you will share all your readings and the densities of your step wedge as read by your Xrite 810 in both modes, we could talk about how significant your results are.
HP5+ test strip made by G-sensitometer channel.
Measured with Xrite 810 (G-channel data), calibrated on Stouffer #21 field 3.05 (my notes say "ISO" CI = 0,49)
0 0,30
1 0,32
2 0,33
3 0,35
4 0,38
5 0,41
6 0,46
7 0,52
8 0,60
9 0,67
10 0,74
11 0,83
12 0,91
13 1,00
14 1,08
15 1,17
16 1,25
17 1,34
18 1,44
19 1,52
Same HP5+ film strip measured with Xrite 361 (V-channel) - (ISO triangle CI = 0,53)
0 ---
1 0,3
2 0,31
3 0,34
4 0,37
5 0,40
6 0,47
7 0,53
8 0,61
9 0,70
10 0,78
11 0,87
12 0,96
13 1,04
14 1,12
15 1,21
16 1,30
17 1,39
18 1,48
19 1,57
After this measurement I measured Stouffer #21 and #8 fields and got 3.07 (vs 3.05) and 1.11 (vs.1,09) in G channel. What I think is rather OK.
---------------------------------------------------------------------------
I think that either my Stouffer is not 3.05 any more in the #21 filed or my 361 calibration target is not 3.01/3.76 in dense fields.
If I calibrate 361T with original 3.76D.REF getting 3.77 (V) on it and then measure my Stoufer #21 field I get 3.17 instead of 3.05D.REF (too big error).
Stoufer #8 field I get 1.18 instead of 1.09D.REF
If I calibrate 810 with Stouffer #21 filed with 3.05 density and then measure 361T target I have 2.92 in G instead of 3.01D (optical/V) or 2.91D (in U-violet mode).REF.
So actually U mode is closer to G channel than V mode?
If I calibrate 810 with 3.01 field from 361T target (instead of densest 3.77 field) using VIS mode (instead of RGB). Reread it as 3.01 (V). Then read #21 and #8 Stouffer's (3.05 and 1.09) in VIS mode and get 3.13 and 1.14.
So looks like both 361 and 810 in VIS mode calibrated by 361t's original target tend to give higher values than mentioned on Stouffer calibrated target #21 and #8 fields.
I don't think that Stouffer could miss so much with the calibration of own target and don't think target could get density with time (I think It could loose density with time).
-----------------------------------------------------------------------------
I have calibrated 810 in VIS mode with 361REF target (field 3.01) and then measured again the same HP5+ strip. (ISO triangle CI = 0,52)
0 0,33
1 0,34
2 0,35
3 0,36
4 0,39
5 0,42
6 0,48
7 0,54
8 0,62
9 0,7
10 0,79
11 0,87
12 0,96
13 1,03
14 1,12
15 1,21
16 1,27
17 1,37
18 1,45
19 1,53
-----------------------------------------------------------------------------
Looks like in VIS mode in density range up to 1.60 both devices 361 and 810 give very close results.
I think I need some new 21 step Stauffer measured/calibrated with trusted 810 device in RGB mode to exclude difference in RGB / Visual modes.
My fog level is 0.30 in RGB mode and 0.33-34 in Visual.