Hi all,


My V30 just arrived from Europe and I’d like to give it a thorough clean before I start using chemicals in it. What’s the best way to properly flush and clean the machine?


Would it be enough to fill it with water and let it run for a few hours, or is there a better process people recommend? There’s still a bit of water sitting in rack 1, and each tank seems to have a small amount of old chemistry left in it. I also noticed some residue in the pipes when I removed the rear and side panels.


I searched through the forum but couldn’t find a clear guide, so I’d really appreciate any advice from people who’ve done this before. I’d hate to damage the machine or ruin my own (or someone else’s) film by cleaning it incorrectly. Thanks in advance!
 
Welcome to Photrio!
I don't know if there are many minilab users on this forum. A few who come to mind, past and/or present, would be @foc and @NortheastPhotographic - I might be forgetting a few others.

I can only comment based on common sense, no hands-on experience, so take this with a grain of salt. Are the hoses easy to remove? If so, I'd be inclined to remove them and clean them outside of the machine. Same for any other parts that are easily removed. Pumps etc. are likely difficult to remove and would have to be flushed in place.

There must have been cleaning agents available (possibly still are) that can be used to rinse the system; might be worthwhile looking into. If you're going to use just water, make sure to replace the water a couple of times so as to dilute whatever muck is being rinsed out.
 
Hi, can I ask a few questions, please, before I give my reply?

Was the V30 bought, - refurbished, - working/as is, - sitting in storage - ?

Is it for personal or commercial use?

Do you have any experience with a minilab/film processor?

( @koraks , thanks for the notification. )
 
Hi the machine was in storage for 2 years but fully working before that, upon starting the machine up for sale the seller noticed there was an issue so they sent it off to be serviced and these were the issues that were found

"What I have checked and done:
F2, the connection for the thermo-sensor was loose causing incorrect temperature.

BL had thermo-sensor error and replaced thermo-sensor.

Float sensor connector was missing/broken off. Found a replacement connector and soldered on.

Replaced the Gas strut for the front cover."

The machine was then tested I guess with water but has now been in storage for about 6 weeks while I figured out what I am doing, while chatting to someone they mentioned now is a great time to give it a deep clean before it starts its new life and thats what I would love to do before putting fresh chemicals in it and starting my developing journey.
 
@biohazard Thanks for getting back.

I would advise you to download the V30 manual,

And read it through so that you are familiar with it.

To clean the machine, first remove all tank racks, keeping the Developing/N1 rack completely separate from the rest of them. Then I would advise filling each empty processing tank ¼ full with clean warm water. Then, using a non-abrasive kitchen scrub (like this)

pot scrub.jpg

and gently clean each side of the tank, and the bottom. Drain and rinse at least twice. For very stubborn stains, I would use methylated spirits/denatured alcohol. This can be used in the pipes/tubing, provided it is well washed out.

I would clean each of the processing racks, out of the machine, in a similar fashion. The N1/developer rack is long, so I would suggest using a kitchen bin/swing bin (50L size) and washing it in that.

When you are happy with your cleaning, I would install all racks, crossover racks and fill with water to tank overflow level. If you have a good supply of circulation filters, then install one full set, and turn on the machine and let it heat up.

When it signals, it has reached temperature. I would put through a leader card on its own.

All going well, it should emerge from the dryer, and all is ok.

Then check your circulation filters. These should be fairly clean, maybe a few block specs. If not, then the pipes and tubing need another cleaning. Maybe remove and clean.

Finally, turn off the machine drain overnight if possible, then mix replenisher (and starter for dev tank), gently fill each tank, with the rack in. Filling sequence should be: Dev, Bleach, Fix, Wash. I would put a plastic cover over the dev tank when filling the bleach tank so as not to contaminate. Then fill replenishment tanks as per instructions.

I would then process a control strip (I assume you have a densitometer) and read and log the control strip. If all is within limits, only then would I shoot a test roll of film and process it.

Can you let us know how you get on, please and how all turned out for you?
 
Last edited:
Hi all,


My V30 just arrived from Europe and I’d like to give it a thorough clean before I start using chemicals in it. What’s the best way to properly flush and clean the machine?


Would it be enough to fill it with water and let it run for a few hours, or is there a better process people recommend? There’s still a bit of water sitting in rack 1, and each tank seems to have a small amount of old chemistry left in it. I also noticed some residue in the pipes when I removed the rear and side panels.


I searched through the forum but couldn’t find a clear guide, so I’d really appreciate any advice from people who’ve done this before. I’d hate to damage the machine or ruin my own (or someone else’s) film by cleaning it incorrectly. Thanks in advance!
I’ve cleaned many. I have never used anything but water to flush the system. I hate the chemicals used by some folks to clean developer tanks etc. My skin is sensitive to something in them.

There are definitely some possible gotchas in the machines.

Disconnect the machine from the mains power before doing any of this. This sounds silly, but some people do try stuff with the power applied. Possible problems are not limited to the machine, but you can get a very nasty, potentially fatal shock. The dryer section is not meant to have fluids in it. It happens and is a safety risk.

1) In the CD plumbing, I’ve seen chunks of plastic jam in the circulation pump. The source of those pieces turned out to be the filter holder. My theory is that the developer was somehow causing the plastic to deteriorate. When folks changed the filter, the torque applied to the bottom lock, caused these small support pieces to break off. Whenever those got into the pump impeller it could stop. Fortunately, cleaning it out gets the pump running again. Note that I never saw this happen in any tank other than the developer. C-41, C-41SM or even RA4, RA-2SM for the paper processors.

(Now that I wrote that, I think it applied more to the V-50 and V-100. I remember the differences between the filters, but not the holder. It might not be the same)

It’s not a bad idea to take the pump apart anyway to clean it out. Poor CD circulation is not a good thing in these machines. Temperature fluctuation and temp. errors can be the result.

2) There’s a CD “shower pipe” that should be cleaned out. It can be removed by loosening a screw near the top of the tank and then twisting the shower pipe unit. It is friction fit. This gets tiny film pieces in it, reducing the flow through the holes. Developer blows towards the film to help circulate better and bring fresh CD in to help strip off byproducts of development. Those film chips are brought in through the front end when the film is cut right through the sprockets.

3) One ugly problem is caused by a combination of low utilization and bleach over concentration. The BL can solidify in the plumbing. Hot water and patience works. I once went on a tour of labs at a particular drug store chain to fix machines with that issue. I was advised to bring overalls and a shop vac. Good advice!

4) Bio growth in the STB was often an issue. I used the water and scrubbing/circulating method here as well.

5) Learn how to disassemble the racks to clean them out. The racks need to be re-timed if the gears are removed. If not, your leader cards will be damaged.

6) Learn how to remove the loading box. Clean it out thoroughly. Remove the sensor bars and clean under them. Many people splash the CD when doing rack maintenance.

Once all that is done, I would fill and dump the tanks several times, circulating in between. Don’t heat the water above operating temperature prior to turning it on. If it’s too hot, you’ll get temperature alarm messages until it cools. Much easier to start low and let the machine warm it up.

The replenishment system needs to be cleaned out and thoroughly flushed. Look for struggling pumps or bellows that “skootch out to one side instead of going straight up and down. I don’t know what you can find for parts these days. I had it easy. :cool:

Bottom line, none of this is difficult. My rule of thumb was not to use water any hotter than my bare hand could stand. Very hot water can distort plastic parts.

Don’t put non Noritsu greases on the gears. Even then, take care to not get any greases on the tank or other structure of the machine. There are meant to be warning stickers on the machine, but nobody reads those. Either that, or they like to gamble.

I don’t remember any crumbling, or other catastrophic tank/frame failures on a V-30, but I’ve seen that on other machines made of the same material.

Whew!