ok...so I cranked up the rpm's to "p", changed my temp +.1 c, used a two step prewash to make sure I was up to 100 degrees f. (total 5 minutes), and added 30" to both the bleach stage and the fix stage.
5'...two step prewash
3:15...Flexicolor dev.
7'....Bleach iii
7'....Flexicolor fixer and replenisher (mixed for the working tank solution, not replenisher solution)
6'....Wash
1.5'....final rinse
The control strips from this recent batch yielded the most exaggerated example of the trends that I've been noticing all along....
Dmin -.03, -.04, .00
LD -.01, .03, -.02
HD-LD .06, .27, .19
Dmaxb-Yb .23
What's weird is that my control strips for the small 2 reel 35mm. tank are much more on point. Same chem., same batch of control strips, same drkrm., same speed on jobo ("p") and same batty operator. Check it out...
Dmin .01, -.03, .01
LD .01, .02, .01
HD-LD .02, .07, .02
Dmaxb-yb .03
Based on these results I'm thinking two things...the 3000 series drum is causing more agitation than the 2500 two reel and thus causing a spike in the green and blue readings (don't they move the fastest when over-developing?)....also, maybe the fixer working tank solution is too dilute for rotary processing, being approx. 85% the strength of the replenisher solution. IDK...maybe there's a problem w/ my rotary speeds, either too fast or too slow. I'll try it at a slower speed later on to compare results.
PE...is there a simple technique to check pH on my bleach? Also, not having any 2% acetic acid will Kodak indicator stop work ok?
Thanks for your help....I'd feel completely in the woods w/out this forum!