I'm seeing slight differences in exposure from print to print. These are prints done of the same negative within a few minutes of each other. I'm using an Omega D2V with a Besseler electronic timer. I do use a timer when developing the print and my agitation in the tray is pretty consistent. Any suggestions as to where I should start in trying to track down the cause? I'm assuming the only potential causes are voltage variations, a bad bulb or the timer. Is there anything that I'm missing or one of these that seems the most likely?

Thanks!
 

Alan9940

Member
Joined
Jun 17, 2006
Messages
2,421
Location
Arizona
Format
Multi Format
What type of enlarger head are you using? Cold light heads are notorious for light output variation due to head temperature. I don't have any direct experience with color heads or condenser heads, but like you I'd suspect voltage fluctuations. Trying to pinpoint these types of problems can be problematic, but your best bet is to pick one variable at a time and eliminate. Good luck.
 

Ian Grant

Subscriber
Joined
Aug 2, 2004
Messages
23,262
Location
West Midland
Format
Multi Format
What about developer temperature, that can vary, I use a dish warme. My main enlargers all have colour heads so voltage stabilised power transformers, that does make a difference.

Ian
 

Anon Ymous

Member
Joined
Feb 7, 2008
Messages
3,661
Location
Greece
Format
35mm
I used to have a Jobo timer. It was electronic and at first everything was fine. From some point on, I had inconsistent print density. I'd make my test strips and the final print would be considerably different in some cases. Eventually, things got worse and it would turn on the lamp for a randomish amount of time. I never really found out what exactly the problem was, but I suspect it was the selector(s) switches. IIRC, they select different resistors and vary the time this way. Perhaps the contacts were problematic, who knows, I really hated the damned thing.

Anyway, see if you can try another timer.
 

Kino

Subscriber
Joined
Jan 20, 2006
Messages
7,757
Location
Orange, Virginia
Format
Multi Format
I knew an old man who had a very old cold light head; maybe an adapted Graflex type? Anyway, he modified a large, electromagnetically fired packard shutter to sit between it and the lens bellows. He kept the cold light on constantly and had the timer wired to the shutter. Claimed he had very repeatable results...
 

Michael W

Member
Joined
Feb 11, 2005
Messages
1,594
Location
Sydney
Format
Multi Format
I'm also thinking voltage variation as a possibility. I've known photographers who use voltage regulators to avoid that.
 

Ozxplorer

Member
Joined
Aug 29, 2004
Messages
229
Location
Gold Coast, Australia
Format
Multi Format
I'm also thinking voltage variation as a possibility. I've known photographers who use voltage regulators to avoid that.

Agreed... voltage stabilisation is a must! My enlargers plus main safelight are connected but not other safelights which I actually see brighten then dim whilst on - an intermittent flickering. However, I’m not sure that these fluctuations can cause an appreciable and visible print density variation as measured in stops or part thereof. Knowing the electrical supply does cause voltage variation I think stabilisation is a wise precaution as it eliminates a variable in the quest for consistency.
 

petrk

Subscriber
Joined
Nov 30, 2017
Messages
119
Location
Prague
Format
Multi Format
I had my exposures off by about stop (I guess) between making test strip print and the corresponding print within about 3 minutes :cry:. Measured variation is ususaly between 220-235V. First cause is the electrical heating in my darkroom, the second is the house heating. There is also variation in the distribution net. What I do now is turning off both heatings which fortunately I can do from my darkroom, so i do it only when working with enlarger. I know I will need the stabiliser.
 

guangong

Member
Joined
Sep 10, 2009
Messages
3,589
Format
Medium Format
My Besseler 45 condenser enlarger came equipped with a voltage regulator, so voltage changes must be a common problem for manufacturer to include as a given.
 

Ian C

Member
Joined
Feb 8, 2009
Messages
1,250
Format
Large Format
0ne way to accurately test your enlarger timer is to connect a Graylab 300 timer to the enlarger timer’s output socket. Preset the Graylab to a longer time than wanted. Then, when you trigger the timer it will run the Graylab for the duration and record the exposure time delivered by the enlarger timer as elapsed time on the Graylab.

Repeat this several times to learn if the enlarger timer is accurate or not and whether the delivered times are consistent or not. I did this to confirm that some capacitor-controlled timers were failing and needed their electrolytic capacitors replaced.
 

petrk

Subscriber
Joined
Nov 30, 2017
Messages
119
Location
Prague
Format
Multi Format
My Besseler 45 condenser enlarger came equipped with a voltage regulator, so voltage changes must be a common problem for manufacturer to include as a given.
Mine transformer is from manufacturer and is marked as stabilised, but its output also varies ...
 
Joined
Apr 14, 2016
Messages
814
Location
Bavaria, Germany
Format
Medium Format
I´ve hat a similar problem twice; once the halogen bulb was just before spent, the second time the stabilizer of a stabilized power source was about to start smoking.

Furthermore, I make sure the system has run for about a minute before measuring with the Heiland Splitgrade, I have experienced readings being off when measuring immediately after starting a printing session. But that really is the first minute only.
 
OP
OP
Dmosher

Dmosher

Member
Joined
Nov 12, 2015
Messages
53
Location
Central Illinois
Format
Medium Format
Thanks for the replies everyone. To clarify, the head is the standard D2V with a 75 watt bulb. Development is within a few minutes so I don't think developer temperature could be an issue and I'm only printing B&W.

I do have several other timers so I'm going to try a different timer first since it seems to be the easiest. I also like the idea of hooking a Gralab 300 timer up to the Besseler timer to check it. I have a couple of those laying around... I like the Besseler timer's operation so I hope that isn't actually the problem. The audible beeping making burning and dodging easy.

Thanks!
 

MattKing

Moderator
Moderator
Joined
Apr 24, 2005
Messages
52,857
Location
Delta, BC Canada
Format
Medium Format
Mine transformer is from manufacturer and is marked as stabilised, but its output also varies ...
Your transformer may be the problem. If it isn't stabilizing, it might be de-stabilizing.
 