Conclusion:
The color of sensitometer light does not affect Tri-X response to changes in development time.
Note that the exposure with the EG&G gives higher speed due to its being "white" light where the LED types are not truly white. Therefore, part of the energy is missing which leads to slower results.
The LED sensitometers are good but will give you what amounts to slower speed.
PE
Thanks for viewing the thread.
Don't read anything in to x-axis variables on those graphs. The x values are just "numbers" or "units" not Lux-seconds. The sensitometers all have different intensities and exposure times and I'd consider the relative right-left position of each curve on the x-axis in the graphs to be random. This was not a film speed test. (That is to come, stay tuned!)
I fully recognize that the intensity of these instruments may be far different. I should have made that post of mine conditional in that respect. Sorry.
If you have any degree of congruence in intensity at one selected wavelength or group of wavelengths, then if the speeds differ in the exposed films, then the spectral output of the group of instruments varies.
Since we see that here, then either the absolute intensity of each device varies or the spectral output varies.
PE
Do you have a similar, prior experience with a Kodak Process Control Sensitometer? I'm not familiar with any of those, but looking for one.
ic-racer,
Not to put you to more work, but can you do a quick Reciprocity study using the EG&G at different exposure times?
I think it would be valuable to show how different the gradient might be from a home-made "flash" sensitometer used at low light output (for example time might be 1/10,000 second at lowest setting).
I'd do the test but I don't have the acetate filters to balance the different speeds, and I think you made a set...
Do you have a similar, prior experience with a Kodak Process Control Sensitometer? I'm not familiar with any of those, but looking for one.
Should be easy to do. I think it will turn out OK. The Wejex in the test had an exposure time of 1 sec and the EG&G was set to 1/1000. There was not detectable difference in the way the exposed film responded to development time changes between those two. So, I suspect 1/10,000 would be similar.
You were on 10^-3? (I usually use 10^-2).