Hi,

As the new gear is coming, it has to be calibrated. This time, I'd like it to be super precise - my current settings is not as good as I would like to be.

The equipment set:

Laborator 1200 with condenser head and Ilford MG filter set
Reflective/Transmissive densitometer
Stouffer 4x5 31 steps projection scale (TP 4x5 31)

I read the procedure in manual and I understand that I should make 7 prints of a scale (the question - should I contact print them, or put a scale in a neg carriage and print on a paper?), develop, measure density of each bar and put step being 1/3 stop number on X axis, and density on logarithmic Y axis getting nice curves. This looks pretty clear for me till that moment.

It is not clear how should I process this data to get contrast value and offset that I will then put into the analyser.

Please advise

Kal800
 

removedacct3

Member
Joined
Mar 19, 2009
Messages
628
Location
-
Format
Multi Format
Hi!

If you would like to calibrate using the step wedge, then you should contact print the step wedge. The exposure time is not relevant as long as you see paper white and pitch black on a single sheet of paper. If that is the case you should measure each bar with your densitometer and do the calculations as described in the calibration manual. Make sure that your head is as high as possible to even out illumination densities.

If you do your calibration like this and you pay attention to your consistency you will get good results.

However, if you would like to improve your results you should read section 5.4 of the ZoneMaster II calibration manual: "Refining the Contrast Calibration". (I am aware that you would like to calibrate your Analyzer Pro, but I believe the exposure and contrast calibrations are alike). The entire calibration process will take considerably longer, as you should determine exposure values first with dried paper, before you can start with your contrast calibrations. This method relies on accurate exposure settings.

Please have a look and do not hesitate to make some noise in case of questions ....

 

peoplemerge

Member
Joined
Jun 15, 2014
Messages
401
Location
Hollywood, CA
Format
Medium Format
If you have not already done so, I recommend Beyond the Zone System by Phil Davis for background reading. It took me a couple of reads for it all to sink in.
 

removedacct3

Member
Joined
Mar 19, 2009
Messages
628
Location
-
Format
Multi Format
Last edited:

Joel_L

Subscriber
Joined
Mar 12, 2011
Messages
579
Location
Colorado
Format
Multi Format
I recently ordered the Zonemaster IIm and Stopclock. I think the ZM II is the same as the analyzer in the analyzer pro. I was hoping to get it before the holidays but doesn't look like that will happen. In any case, I will be going through this exercise too.

If you learn anything interesting, please post.
 
OP
OP

kal800

Member
Joined
Jan 12, 2016
Messages
155
Location
Gdynia, Poland
Format
Multi Format
Hi,

I'm struggling to get my Foma FB paper calibrated with condenser enlarger and cannot get any success. Long story short - head all up to the top of the column, lens stopped down do f45, G2 filter, shortest time possible - 1:15 seconds, empty neg carrier in, LED light source of 6500 K, 2450lm. All I can achieve is D0.08, I am not able to get referential D0.04 at all. Anybody ever calibrated Foma paper with condensed head?
 

albada

Subscriber
Joined
Apr 10, 2008
Messages
2,172
Location
Escondido, C
Format
35mm RF
Hi,

I'm struggling to get my Foma FB paper calibrated with condenser enlarger and cannot get any success. Long story short - head all up to the top of the column, lens stopped down do f45, G2 filter, shortest time possible - 1:15 seconds, empty neg carrier in, LED light source of 6500 K, 2450lm. All I can achieve is D0.08, I am not able to get referential D0.04 at all. Anybody ever calibrated Foma paper with condensed head?

I think you're trying to get the lowest density the paper can provide (D-min). If so, have you tried developing a piece of unexposed paper? My tests showed that the D-min of Foma RC is 0.11, so the 0.08 that you're getting for FB seems reasonable for Foma. For comparison, Ilford RC and FB measure at 0.06.

Mark
 

removedacct3

Member
Joined
Mar 19, 2009
Messages
628
Location
-
Format
Multi Format
Hi,

I'm struggling to get my Foma FB paper calibrated with condenser enlarger and cannot get any success. Long story short - head all up to the top of the column, lens stopped down do f45, G2 filter, shortest time possible - 1:15 seconds, empty neg carrier in, LED light source of 6500 K, 2450lm. All I can achieve is D0.08, I am not able to get referential D0.04 at all. Anybody ever calibrated Foma paper with condensed head?

I did successfully calibrate my zonemaster with Fomabrom paper using a LED source, but I needed a ND filter to do so. That would be my advice as well, get an ND filter for your enlarger lens.
 

Ulrich Drolshagen

Subscriber
Joined
Mar 14, 2005
Messages
526
Location
Germany
Format
Medium Format
1:15 seconds
This information seems ambiguous to me. Is it 1.15" or is it 1' 15"?

If it is really 1.15" stopped down with the head all the way up there seems to be too much light to be practically usefull in general. I'd be inclined to rework the lightsource first.

Ulrich
 
OP
OP

kal800

Member
Joined
Jan 12, 2016
Messages
155
Location
Gdynia, Poland
Format
Multi Format
That’s my point - if I opened up the lens two stops and lower the head to have standard enlargements, I would have to install a LF lens with shutter. I’ll start with getting weaker bulb.
 