From memory:
At 20C, the ideal glass pH electrode would have, according to Herr Nernst, a slope of 58.2mV/pH and an intercept of 0mV at pH 7.
The ideal slope drops to something like 54mV/pH at 0C and increases to something near 64mV/pH at 50C AT 40C I'd guess it is around 62mV/pH.
What it all works out to is something like -0.2 mV/C to adjust the slope.
This all assumes a perfect probe which will never be the case. Your best bet for a meter without ATC, is to calibrate at or near the temperature of interest and if you can, using the charts that come with the buffer solutions, enter the corrected pH for the buffer temperature.
If you are using an IsFET or Iridium oxide probe, the idea is the same but the slope and intercepts are different.