Luna Pro exposure meter - please confirm my understanding

chuckroast

Subscriber
Joined
Jun 2, 2023
Messages
2,106
Location
All Over The Place
Format
Multi Format

Not sure what you're saying here.

The MR9 adapter will take 1.5V silver oxide batteries that will bias the diode reasonably well without any huge loss of calibration accuracy.

Moreover, the silver oxide batteries will fail more-or-less suddenly just like a mercury cell.

I have worked on many of these meters and have found that either an internal recalibration or a silver oxide/diode
combination works about equally well. Since silver oxide 625 batteries appear to no longer be available,
for those meters I have recal-ed for 1.5 volts, I just use physical MR9 style adapters without a diode and stick silver oxide 386s or SR44s in them, depending on the application.

There is also someone selling small voltage regulators that can be installed internally in the meter to hold the internal VCC at 2.7 volts, but that does involve tearing them open and some soldering.

Let us keep in mind that these meters were accurate to about 1/2 stop best case and were decidedly nonlinear from the top of the low range to the low end of the high range.

All that is to say that - at least for monochrome, and possibly for color film too - it's close enough. As always, repeatability is more important than absolute accuracy.
 
OP
OP

336v

Member
Joined
Mar 21, 2025
Messages
22
Location
USA
Format
Pinhole
The extreme ends of the scales are somewhat nonlinear on these meters. The top of low range will not perfectly match the bottom of the high range even when correctly calibrated.
It does on mine - amount of light deflecting pointer to "12" on low scale, deflects it to exactly 12 on the high scale after I've calibrated it.
12 is the only common readout for both scales.

Non-linearity within each scale is taken into account by offsetting tick marks on the scales - you will notice that beyond "6" tick marks no longer
coincide, one on top is skipped, and the high range bottom scale looks squished. The same deflection angle has 9 gaps between tick marks
on top, but 10 on the bottom. This graphic "fix" is not strictly ideal, but more than accurate enough for practical photography.

Gossen knew this when designed the thing.
 

BrianShaw

Member
Joined
Nov 30, 2005
Messages
16,428
Location
La-la-land
Format
Multi Format
...

All that is to say that - at least for monochrome, and possibly for color film too - it's close enough. As always, repeatability is more important than absolute accuracy.

It is true for color, including color positive film. I think of it this way: Gossen (and probably all other light meter makers) had experienced and knowledable EEs on their design and manufacturing staffs, plus these meters wouldn't have lasted in teh market as long as they have if they weren't up to the vast majority of photographic tasks. Indeed, they knew what they were doing.
 

Bill Burk

Subscriber
Joined
Feb 9, 2010
Messages
9,162
Format
4x5 Format
Learn Camera Repair site is run by generous people who faithfully gathered manuals. For a long time they were free but they recently changed to a pay-as-you-go site. This is to distinguish them from scam sites that turn up in every Google search result.

I put a 2.7 v regulator from Filip Dee in my Luna Pro and it immediately acted as if calibrated. So I didn’t touch the dials inside.
 
OP
OP

336v

Member
Joined
Mar 21, 2025
Messages
22
Location
USA
Format
Pinhole
Yes, if this curve to be trusted, the SR43 is quite a bit more superior to SR44, so for these battery non-linearity
of the curve can be taken out of discussion. Only the diode's limitations remain. Good find!
 

Dan Fromm

Member
Joined
Mar 23, 2005
Messages
6,799
Format
Multi Format

I have two of the long-discontinued original. Injection molded. 3-d printing may be less expensive and might allow production to order, reducing the risk of being stuck with unsaleable inventory. AFAIK the originals are passive. I've always thought they had a Zener diode. I used one in a LunaPro (= LunaSix III) and the other in a Canon AZ 814. Both worked. Both are now in a Horseman Optical Exposure Computer and still do the necessary.
 

Chan Tran

Subscriber
Joined
May 10, 2006
Messages
6,698
Location
Sachse, TX
Format
35mm
I found that using 2 resistors one in parallel and one in series to the CdS cell and playing with the values of these 2 resistors I can get the current relatively linear to LV (which is logarithmic). It's kind of an S curve.
 
OP
OP

336v

Member
Joined
Mar 21, 2025
Messages
22
Location
USA
Format
Pinhole
I found that using 2 resistors one in parallel and one in series to the CdS cell and playing with the values of these 2 resistors I can get the current relatively linear to LV (which is logarithmic). It's kind of an S curve.

Yes, you can do it, there are ways to re-shape the galv. response electrically, but keep in mind that based on the circuit any addition of either series or parallel resistors to CdS or existing trim pots
will either decrease sensitivity or increase power consumption (or both). The shape of CdS resistance to light response (which is log function, e.g. very straight line on a log scale, not S-curve)
is taken into account when graphic tick marks were silk-screened on the meter's scale plate. The S-curve you're referring to is caused by the decrease of sensitivity of the galvanometer itself near extremes
of deflection, it is the most sensitive in the middle, thus you see black swatches stretched between about 1/3 and 2/3 of deflection angle and squished toward the ends. This got nothing to do with CdS's Lux/R curve. Just keep in mind that loading the CdS with extra resistors, while will linearize needle deflection per each EV step, will invalidate this designed-in "graphical compensation", so overall accuracy near the ends of scale(s) will certainly be worse.
 