Jobo Recirculation Problem

Joined
Mar 8, 2005
Messages
24
Location
Farmington,
Format
Med. Format RF
I recently purchased a Jobo ALT-2 with a water bath recirculation problem. When the power is first turned on, the recirculation pump runs for 5 to 45 seconds or so, and then stops. I have to older ATL-2's that I can use for parts, but am not sure what is happening. Any thoughts?
 

AgX

Member
Joined
Apr 5, 2007
Messages
29,973
Location
Germany
Format
Multi Format
Could a temp. sensor react after only 5sec of assumed over-strain?
 
OP
OP
Joined
Mar 8, 2005
Messages
24
Location
Farmington,
Format
Med. Format RF
Is a circuit-breaker or over-temp sensor tripping or something? Is the water actually circulating with the pump running?

Sorry, I should have been more specific, and provided more info.

Yes, the water is actually circulating. As I said, it works for a few to several seconds and then stops. If I turn off the power, and immediately turn it back on, the water again recirculates for a few to several seconds—then stops.
 

CatLABS

Member
Joined
May 20, 2011
Messages
1,576
Location
MA, USA
Format
Large Format
Even if the thermal protection circuit is tripped, you would have had to reset it before the motor came back on.

This is often caused by debris build up in the pump shaft housing, Or the motor is otherwise obstructed in some way, causing it to heat/seize.
The pump shaft housing is easily removed and cleared.
If that does not do it, you will need to lubricate the pump motor it self. One - two squirts of lube spray (we like PB plaster) will do.
Opening the housing is a bit more challenging but there are details instructions on our manual section
www.jobo-usa.com/resources/manuals

The motor might also have a corroded shaft, which might be beyond repair, you will only find that out when you open it up and have a look.

Good luck.

Let us know what you find.
 