Hi Everyone:
I have received my ATL2 and as I knew when I bought it it is showing an air distribution error.
In diagnosing this problem, I checked for output from the air pump. Nothing. I then disconnected it, and found no 24 volts going into it. I then provided 24 vdc from an outside source directly to the pump, and it does operate.
So the problem appears to be an interruption of the 24 vdc going to the pump. Now, the next step I took was to check the fuses. There are three fuse holders in it Fuse1, Fuse 2, and Fuse 3. Only Fuse 1 and Fuse 3 was marked, and both their fuses are good. Fuse 2 is an unmarked empty fuse holder. So now, it is difficult for me to tell if the fuse 2 holder was merely used to close a blank in the board or whether indeed there is a missing fuse which should have gone in it (and if so, what amperage).
Other functions on the processor do work, e.g. to raise and lower the drum from the front panel switches.
A large part of my problem is my location - I live in Atlantic Canada, and to the best of my knowledge there are no longer any repair sites in Canada for Jobo's. The cost of postage to the US is approximately $100 to the US one way, so it would be impractical to send it to California, so I am going to have to diagnose and fix it myself.
The photo below shows the pump at the left front and behind it, the air distributor..
Does anyone know if the middle fuse is actually used in an ATL1, 2, or 2+ processor, and what value? Even more helpful would be a schematic or wiring diagram for the unit.
Can anyone offer advice?
Dave
I have received my ATL2 and as I knew when I bought it it is showing an air distribution error.
In diagnosing this problem, I checked for output from the air pump. Nothing. I then disconnected it, and found no 24 volts going into it. I then provided 24 vdc from an outside source directly to the pump, and it does operate.
So the problem appears to be an interruption of the 24 vdc going to the pump. Now, the next step I took was to check the fuses. There are three fuse holders in it Fuse1, Fuse 2, and Fuse 3. Only Fuse 1 and Fuse 3 was marked, and both their fuses are good. Fuse 2 is an unmarked empty fuse holder. So now, it is difficult for me to tell if the fuse 2 holder was merely used to close a blank in the board or whether indeed there is a missing fuse which should have gone in it (and if so, what amperage).
Other functions on the processor do work, e.g. to raise and lower the drum from the front panel switches.
A large part of my problem is my location - I live in Atlantic Canada, and to the best of my knowledge there are no longer any repair sites in Canada for Jobo's. The cost of postage to the US is approximately $100 to the US one way, so it would be impractical to send it to California, so I am going to have to diagnose and fix it myself.
The photo below shows the pump at the left front and behind it, the air distributor..
Does anyone know if the middle fuse is actually used in an ATL1, 2, or 2+ processor, and what value? Even more helpful would be a schematic or wiring diagram for the unit.
Can anyone offer advice?
Dave