Stephen Benskin
Member
I haven't noticed anyone mention the ANSI tolerances which are on the order of +/- 25%.
When I was doing repairs we used a standard light source built into a Kyoritsu tester.
It used transmitted light not reflected and in addition to the brightness being variable there was a "K" factor that would also change the value perhaps by slightly changing the spectrum?
Anyway if you have two meters differing by 1/2 stop, each of them could be within tolerance but at different ends of the scale. If you want them to agree you would need to send them in at the same time & have them calibrated to each other.
K is just a light loss constant that factors in such variables as the light loss as it passes through the lens and the accuracy of the exposure meter. There's a fairly large equation in the standard to calculate K, and in the 1971 standard's appendix, it recommends removing the most changeable variables from the equation to create a stable value of K, they label as Ko. This saves you the trouble of working with too many numbers. This way you are only working with the two that change most often, and they are the actual transmittance of a lens and, most importantly, the spectral response of the photo cell in the meter.
The idea of K is that when the meter is aimed at the calibrating light source and it doesn't read the light source correctly, you add a value that makes up the difference which is called K.
According to the most recent standard, there are two ranges of K. Each has a range of +- 1/6 stop.
Mike, a T.V. remote should do the trick. I just checked mine and found a 1/3 to 2/3 stop change in the meter's response.