I got a Heiland densitometer and love it.It comes with a long cable so I can read densities on large prints.In 20 years, it never needed an adjustment or maintenance.perfect equipment; made in Germany.I am thinking about picking up a densitometer for my film calibration. Any advice on a good value that will read RGB as well as regular readings? I would also appreciate any "beware of" advice. Don't want to waste my money on a POS. Thanks in advance.
Questions to X-rite Owners (810, 811, 820)
How do you calibrate color transmission (if you do it at all)? X-rite no longer sells transmission calibration film. However they do sell a reflection plaque.
The problem is that with 811, you have to calibrate transmission first before you calibrate reflection according to an x-rite support article.
Of course one can calibrate with stouffer wedge for visual but then your RGB will not be in calibration.
I think ParkerSmithPhoto just got a densitometer I recommended...
Now to see if it is all I say it is...
Use S-11 clear bulb it says on the bottom...
wattage? .... depends on the intensity of the lightning strike and length of the kite string.
I picked up a couple to try from my local hardware store.
Hi Mr Bill,Bill, all of the transmission densitometers I've ever worked with directly, prior to the automated strip readers, used internal optics to project an image of the lamp filament into the measuring aperture. This relies on having a specific filament shape that has to be aligned properly when the lamp is replaced. Whenever we had a densitometer that kept jumping out of calibration, it was almost always due to a lamp with a damaged filament; the aperture was only partially filled (with the image), and using the densitometer caused the filament image to jiggle; periodically settling into different positions.
I'm guessing that some random bulb is not going to be right? Assuming that your densitometer works on the same principle. Anyway, it's something to be aware of as a possible issue.
If I throw out two lowest and four highest readings where the greatest deviations occur, I get standard deviation of 0.022 with the original bulb and 0.019 with the new bulb.