the results when scanning are good despite what the strips say.
I use the Fuji strips. The readings are spot-on (unless I mess up temperature). Feels like your chemicals may be an issue here... Are you using blix-based kits? Or "old" Flexicolor? I am using Bellini C-41 or Fuji C-41 5L from Freestyle.
Scanning is noticeably easier when readings are on target. Specifically, inverting colors is easier.
Is it correct you're running a relatively low-volume setup, doing C41 in batches of a few films at a time, in tanks (rotation or manual inversion)?
What is the volume of the working stock of your developer, how are you replenishing it, and how do you store the working stock?
What are the deviations you generally see on your control strips?
I see you're adding components of the developer instead of replenishing it with the regular replenisher. Is this based on some technical publication of Fuji?
Alright, I see.
I suspect that it'll be hard to get things stable in (1) a relatively small volume, (2) if you dump everything after a few days and start over and (3) if you throw things off-balance with haphazard interventions. I'd be inclined to give it some time to settle, and only replenish using the actual replenisher. If you see high densities increase over the target value, cut back on replenishment. If they drop, replenish a little more. Then hope that things will even out.
I'd also suggest getting in touch with FUJIFILM and discussing your problem with them. If your chemistry is FUJIFILM Hunt, contact the people in Sint Niklaas, Belgium. If it's manufactured somewhere else, try through the US FUJIFILM offices and see if they can get you in touch with someone who can advise you in this.
PS: I assume replenishment on your machine is set in accordance with the spec sheet of the chemistry you're using.
if you know other people it would be amazing to know
Valid points.Are you sure all of your circulation pumps are working properly ? ... with good clean filters ?
Are you sure all of your circulation pumps are working properly ? ... with good clean filters ?
If you're processing 200+ rolls ( with relatively small tanks ), then there's something seriously wrong going on here.
John S
1) would you mind elaborating on what problems dumping my working tank can be introducing?Please stop dumping your working tank solutions, this is causing more problems than it is solving.
I have a few questions, if you don't mind, about your workflow.
As @koraks has mentioned above, there are too many variables happening to be able to pin this down, when it should be a fairly simple job.
- What densitometer are you using?
- Do you keep the control strips in a fridge (chilled)?
- When you take them out do you let them come to room temperature?
- Do you use an instant read thermometer to check the developer temperature daily and before control strip processing and do you log these temperature readings?
- Do you measure the reference strip and apply the correction factors supplied with the control strips?
- On the Noritsu V30, do you check your replenisher pump out, and circulation filters on a weekly basis?
- Has the Noritsu V30 been serviced and calibrated recently?
You need to let a freshly mixed developer tank working solution "settle in". I would suggest that with any freshly mixed tank, make your first control strip reading and then it should be let run for at least a week before another control strip readings is made. With 7.5 liter tank of fresh developer, there is plenty of active developer and replenisher and a tank that size shouldn't go bad that quickly.
The secret to maintaining good quality control is to make any small adjustments one at a time. And if you go slightly outside the control limits, it's not the end of the world, it can be brought back in. If it keeps going outside the limits then there is a problem.
If you are processing low volume of films, then I suggest you read this.
Fuji C41 low volume
In the Fuji document, see this chart to help select the best low volume option for you. The V30 has 7.6 Liter developer tank so look at the 10 liters in the chart.
View attachment 368690
I do, but I doubt they have the time to respond to this, and their first suggestion would most likely be what I already told you. I don't think they're very likely to respond to a problem that boils down to the combination of "a multitude of factors and interventions" resulting in "erratic density shifts in all directions and at all levels" - or, in a single word: "chaos". This is not to discourage you, but to emphasize what @foc says above as well: isolate factors, allow for settling time, small incremental changes. Then, once you get a clear pattern to the drifts you're seeing, feed that to Fuji and see what they say. But in all likelihood, the problem will have resolved itself by then...
Valid points.
one thing I'm noticing from both you and Koraks is you guys are stuck on my replenishment etc.
1) would you mind elaborating on what problems dumping my working tank can be introducing?
2) Xright 810 - calibrated
3) yes they are stored cold
4) sometimes I don't wait as long as I could. I'll run a new set of test strips with all fresh chemicals and see what I get nd report back here
5) I do not record my CD temperature and log it. What benefits will I gain from doing so?
6) I do. I follow kodaks procedure on z131 with fuji's control strips and graph everything on a kodak graph
7) I keep awareness of the replenisher pumps, and I've uploaded a video of the CD circulation pump (all the others are fine, but please let me know if you think the CD is fine in the video. (read reply above for link to video)
8) it has not been serviced or calibrated since I purchased it a year ago in very good condition
MY TEST STRIPS HAVE NEVER BEEN IN ORDER SINCE I SWITCHED TO FUJI. EVEN WHEN MIXING NEW CHEMICALS, MY PLOTTINGS ARE OFF THE CHARTS. INFACT THE CORRECTION FACTORS MAKE EVERYTHING WORSE - DESPITE THIS THE FILM COMES OUT VERY NICELY. GREAT CONTRAST, VIBRANCE, AND COLORS
I hear what you say about your consistent problems, but doing odd stuff with replenishment isn't going to make it better. Two wrongs don't make a right.
Besides, the information you're feeding us is (1) you have trouble with strip consistency and (2) you're doing odd things with your chemistry. It's only fair we point this out. I do accept that given what you state, there is probably more going on. It's always possible you somehow made a mixing error with the chemistry or you're working with an expired batch (although the factory packaging has reasonably good shelf life in my experience). Or there's some misc issue with contamination, temp variations in your machine etc. etc. It's a bit of a crap shoot at the moment and most of that is not your fault; it's just challenging to get a grip on what's going wrong with someone's setup a couple of thousand miles away. Heck, it's challenging enough standing next to the machine, as you've noticed.
If you mixed the developer tank solution correctly (replenisher, starter & water) then there should be no need to dump. Remember what I said about letting the tank solution "settle". Am I correct in that the negatives look and scan OK but the densitometer reading is off and that's why you dump the chemicals?
The benefit is that you have a record of the temperature of the developer over time. Information is always helpful and if consistent (when viewed on a chart or log) can eliminate temperature as a cause of a problem.
I would recommend a weekly log of control strips, developer temperature readings, all tanks replenishment rates and pump output.
I think I filled it with water so that the machine wouldn't say "low tank"Circulation looks good. The tank looks full even with the rack taken out. Normally it should look just over half full with the rack out. Also the machine should alarm when the rack is out and the levels drop. The circulation should then stop and heater cut out. Did it do this?
Am I viewing that correctly and why is it looking full?
meaning I should contact noritsu directly? on on the east coast of the USA, perhaps you know some good people I can go to?I would recommend that the machine should be serviced at least once a year by a Noritsu engineer (if possible).
not weekly no, but I've done it many timesI think your problem may be here. I think something is causing incorrect readings.
Can you get another densitometer, borrow from someone, and compare readings? Or could you send your control strips to someone that can make reliable readings for you. (Fuji in Ireland offered this service years ago)
The X-Rite 810 recommends you do the transmission calibration weekly. Do you do this?
Is the opal glass cleaned regularly and the optic blown with compressed air to free any dust?
I've never heard of these chemicals before. These are the chemicals I know of (attached some screenshots) - and when I google these they do come up, but I'm a little hesitant. do you happen to know about the difference between these two lines of chemicals?Reading the chart: First how many films (on average) do you process?
View attachment 368745
Your Noritsu V30 has a 7.6 Liter tank so look at the row 10L. At the bottom is the number of films processed per day. I have high lighted in blue in the chart.
If you process 10 films a day (on average) then you should be using (yellow 1) EnviroNeg Developer Replenisher 60AC @ 60ml per 35mm 24 exp film.
If you process 15 films a day (on average) then you should be using (green 2) EnviroNeg Developer Replenisher AC.
One last note. Do you remove the cross over racks, in sequence, every night and put them in a water bath (a 10x8 print tray filled with water or something similar)?
Do you leave the top cover/lid of the machine slightly open/vented to stop condensation from forming on the inside of the lid?
I hope this helps. It would be great if you could get someone local to you, with the relevant knowledge, to help.
Be prepared to pay for this knowledge. Ideally a Noritsu C41 service engineer, they would be worth their weight in gold and probably as rare as hen's teeth.
I'm thinking the control strips have been compromised.
Do you have a friendly 'local competitor' ( or somebody on here reading this ) who could run a strip in their processor ?
Did you buy the strips from overseas ..... with possible x-ray damage ????
How does your processed strips look visually next to the reference strip ?
John S
since the machine has temp control I'll run off the assumption that the temperature is always in control until I see a drastic enough change in control strips. I don't see why I should constantly assume the temp of the machine is off
I'll bring my control strips to another lab and see if they get the same results
meaning I should contact noritsu directly? on on the east coast of the USA, perhaps you know some good people I can go to?
update my processor broke so I have to wait till it's all fixed up before I can do anything further
Ouch, sorry to hear that.
Is it a defect that may be related to the variable results?
Sorry for your troubles. What broke?
Are you repairing yourself or an engineer?
If an engineer, ask them about your process control problems and see what they say.