I've seen this question posted on APUG before and I still don't have the answer. I bought a used X-Rite 810 densitometer. It comes with a reflection calibration tablet, but it's missing a transmission step wedge. Does anybody know the densities of the particular patches on the step wedge? I looked at the manuals online and they don't specify the densities. Anything piece of information will be helpful. Thanks in advance
 

bascom49

Member
Joined
Oct 14, 2010
Messages
231
Format
Medium Format
I have the same densitometer as you.
You can order new ones from x-rite. They are not very expensive as I recall.
 

Chan Tran

Subscriber
Joined
May 10, 2006
Messages
6,840
Location
Sachse, TX
Format
35mm

Each wedge is different slightly form each other. You set the 810 for the values on the wedge. The step that is used to calibrate (CAL high) is about 3.00 density. The CAL low is nothing. There are steps of 0.27, 1.50 and 3.75 to check for linearity.
 
OP
OP
Joined
Dec 10, 2009
Messages
6,297
Format
Multi Format
Thanks Chan!

Each wedge is different slightly form each other. You set the 810 for the values on the wedge. The step that is used to calibrate (CAL high) is about 3.00 density. The CAL low is nothing. There are steps of 0.27, 1.50 and 3.75 to check for linearity.

I have a Stouffer step wedge with known densities and I'll use it. I assume I can set the 0.27, 1.50 and 3.7 densities with my step wedge with similar densities also.

Thanks again!
 

Chan Tran

Subscriber
Joined
May 10, 2006
Messages
6,840
Location
Sachse, TX
Format
35mm
I have a Stouffer step wedge with known densities and I'll use it. I assume I can set the 0.27, 1.50 and 3.7 densities with my step wedge with similar densities also.

Thanks again!

Yes! the other steps can be any values as they are used only to check if it reads correctly.
 