I recently received an X-Rite 310T. It arrived without the calibration wedge, but I have a film of approx. D=3 to calibrate it. My question is: Should the 310T retain calibration values when switched off, but left plugged in, or does it have to be recalibrated each time it is turned on (and allowed to warm up for a few minutes)? When switched off it looses the calibration information. It retains the time/date information for a few hours, but the clock stops when turned off, and resumes when it is powered up again. The on-line instructions say that it should be recalibrated once a week. Does this mean it has to be keep on 24/7 (seems it would burn out the lamp pretty fast)?
Thank you.
Clarence
Check out this link
http://www.xrite.com/product_overview.aspx?ID=24&Action=support&SupportID=14
It would retain the calibration after power down.
You film with a density of 3 is too high to calibrate. The high density in the xrite calibration wedge is about 1.50.
I recently received an X-Rite 310T. It arrived without the calibration wedge, but I have a film of approx. D=3 to calibrate it. ....Clarence
Hi Clarence, how accurate is your calibration reference of approx D=3 ?
The whole point of having a calibration reference is that it is at least 4 times more accurate (and preferably 10 times more) than the instrument you are calibrating with it.
regards
Peter
Clarence! I am sorry and you're right. I have 2 of those 319-68 Xrite step wedges. And yes Cal high are around 3.00 density. (they are a little different on each wedge). The step 2 which is for linearity check is around 1.50.
I recently received an X-Rite 310T. It arrived without the calibration wedge, but I have a film of approx. D=3 to calibrate it. My question is: Should the 310T retain calibration values when switched off, but left plugged in, or does it have to be recalibrated each time it is turned on (and allowed to warm up for a few minutes)? When switched off it looses the calibration information. It retains the time/date information for a few hours, but the clock stops when turned off, and resumes when it is powered up again. The on-line instructions say that it should be recalibrated once a week. Does this mean it has to be keep on 24/7 (seems it would burn out the lamp pretty fast)?
Thank you.
Clarence
Inexpensive calibration step wedges can be purchased from Stouffer Industries.
http://www.stouffer.net/TransPage.htm
Charlie, I have had a couple of the Xrites with dead batteries in the past. I found an aftermarket battery...albeit I needed to solder the required tabs to the battery and then into the board...but it works fine and the cost is less, if that interests you.
As I recall, the last battery was found at a local Walmart store...not sure if they have those up where you live.
Hello PhotoSmith,
Sorry to have not said this in the last post.
Welcome to APUG! I hadn't noticed that it was your first post.
Cheers,
Clarence
We use cookies and similar technologies for the following purposes:
Do you accept cookies and these technologies?
We use cookies and similar technologies for the following purposes:
Do you accept cookies and these technologies?