| Densitometer | Emulsion Direction | Is it in the manual? | Notes |
| X-Rite 810 | Facing down | No | 810 and 820 are the same optical mechanism, polished opal diffuser and sensor elements in the base |
| X-Rite 820 | Facing down | No | |
| X-Rite 361T | Facing up | No | |
| X-Rite 301 | Facing up | Yes | |
| Tobias TBX1000C | Facing up | Yes | Polished opal diffuser and sensor elements in the head |
| Heiland TRD-2 | Facing down | Yes | |
| Dektronics Printalyzer Densitometer | Facing up | Yes | Acrylic diffuser in the base, sensor in the head, but calibrated to compensate for measurement artifacts. |
| Dektronics Printalyzer UV/VIS Densitometer | Facing down | Yes | Polished opal diffuser in the base, sensor elements in the head. |
Warning - pedantry alert!
Don't you mean emulsion orientation, rather than emulsion direction?
To me, direction implies movement.
Interesting. That might explain some odd behaviour with my old Photronix Delta III (light in the base, sensor in the arm above). Needless to say, it came my way years ago without instructions!
Pedantry accepted, within reason.
Thread subject changed, but I didn't upload new graph screenshots.
Pedantry accepted, within reason.
Thread subject changed, but I didn't upload new graph screenshots.
What you meant was obvious to everyone except the pedant.
And the moment you try comparing readings across densitometers, or across lab instruments capable of measuring transmission density (e.g. spectrometers), you'll quickly notice that the nitpicky peculiarities of light diffusion make a very noticeable difference in how well all of your results match up with each other.Worth noting Todd-Zakia in Photographic Sensitometry names three types of density: 1. Specular; 2. Diffuse; 3. Doubly Diffuse
and three further classifications related to the nature of the receptor: 1. Visual; 2. Photoelectric; 3. Printing
And the moment you try comparing readings across densitometers, or across lab instruments capable of measuring transmission density (e.g. spectrometers), you'll quickly notice that the nitpicky peculiarities of light diffusion make a very noticeable difference in how well all of your results match up with each other.
Most desktop densitometers are diffuse (either influx or efflux). I think the setup that NIST uses for the 38120C (a.k.a. SRM 1008) might be doubly diffuse, but its not clearly stated . They have a whole document (NIST SP 250-99) covering the specifics of their setup in excruciating detail, and that was my takeaway from some of the figures.
Ultimately, the setup used for "the master standard" and the setup used by your actual densitometer might not match perfectly. But if things are calibrated correctly, and you're generally measuring the same sorts of materials, it might not matter.
The biggest issue is when you do things like mismatching film orientation between calibration and measurement. In that case, you get this low density "hump" in the results.