So leave your Kodak thermometer in the water bath and tweak the Jobo temperature knob until you are happy.
It will take some time for the temperature to stabilize. Maybe put in a tank and let it roll to swish things about. And also watch to see how much hysteresis the Jobo controller allows.
I've only ever done B&W with mine. I don't worry much about the absolute temperature since I'm going to be testing for personal EI and developing time anyway. Just always set it to 20C and wait to start a run until I hear the relay cycle a few times.
And that is why I suggested letting a drum spin to splash the water around a bit and reduce the temperature gradient...I have both a CPE2+ and a CPA2 and to get 38 C in the chemistry, I must set the JOBO's to 37 C.
To add to the confusion, try measuring the temperature of the JOBO water bath on different places...
Harry, you live like a risky man. Bring some of your slides on the next meeting, will you? I'll bring mine.
G
That indicates that you have two thermometers that agree closely. Whether or not those points of agreement is, are fact, the CORRECT temperature, appears to be up for grabs.Based on suggestions made I checked the calibration of my Kodak Process thermometer against a medical thermometer (the old fashioned max reading type). The two agreed exactly.
The E-6 1st & color developer require 38°C ± .3°C