You can probably still buy calibration plaques, but they may cost more than your densitometer did - a reflection calibration plaque for an Xrite 810 or 811, etc, ought to work.
The normal calibration routine is that you would put your probe down on a white calibration target, while pressing a zero button. Obviously the white plaque is not supposed to read zero, but the aim has to be adjustable somehow (perhaps there is a "zero" knob on the back of the machine?). Then while reading a dark calibration patch, adjust a slope control (perhaps another knob, labeled "slope" or "calibration"?). Assuming that this is a color machine, you would have to zero color, individually. I'm thinking that you can also set slopes, individually, but not sure. Does your machine have controls like this? Note that there will be a limit to adjustability, so you might want to try a test to see what the ranges are.
The machine has cables for a printer and a computer...
Anyone Know anthing more about these two uses?
Software/Hardwaren needs?
Sorry, my manual is for tjhe TD-504. My oops.
Holy cow, Mike, that's an oldie! Probably nixie tubes in the display, too, right?
there's not much practical value - you could write down the numbers nearly as fast.
Regarding the manual, it's probably not worth spending much time on a search. If you can calibrate, and know what the colored buttons on the turret are, that's about it. Plus how to change the lamp and any internal fuses.
Hope you can get it all working and get some use out of it!
Humm, well there is a problem with the reflection reading head turret...
as is typical with these Macbeths, the polka dotted disk has turned into a UFO and is now MIA.
Should be a piece of cake to identify the "who is what"... find a few color samples to read: red, green, and blue. Not too critical, photos with a bright red ball cap, green leaves, etc. Reading the red object, whichever turret position gives the LOWEST DENSITY is the red-filter position, likewise green test target for green-filter position, and same for blue. The remaining turret position is the visual filter. If there's any question about the blue filter (a lot of people would identify a cyan test color as blue), try reading a yellow test patch - this time the HIGHEST density reading identifies the turret's blue-filter position.
Mr. Bill,
Thanks for going to the trouble.
I pulled out my MCC today;Was going toi do today, butgot sidetracked when I found software fore putting the measurments into excell, but discovered my cable seems not to be the RS232 Dsub25 nor the Dsub 9, but an HRS (Hirose) RC40-24PR for which I have no idea what to do with....
...
Might be IEEE-448. http://en.wikipedia.org/wiki/IEEE-488
Ah, yes. The good old GPIB interface. As a parallel bus, it's not directly convertible to serial...A better bet is to look for a used card out on one of the Internet dumping grounds for stuff (i.e., 'the bay'). If you have a card, NI is pretty good with support.
Ed