Marco B said:
(snip)
Can I safely assume, taking into account the fact that I have set the 0-reference of the measurement instruments in a similar manner using an unexposed clear piece of film, and assuming well calibrated instruments, that a density value of X means a similar film density on instrument A as compared to B?
E.g. Does a film density value of 0.82 measured on machine A and B represent the same actual film density?
Thanks for responses.
Marco, it depends upon how exacting you wish to be; to how many decimal points.
Each densitometer, within a brand as well as with brand/component combinations, can vary in linearity and accuracy depending upon user calibration and on baseline calibration of the system to traceable standards.
If you are hoping to have two densitometers agree 100% on all measured film densities, I don't think you are going to be happy at all unless these units are of the same brand, model, very close to each other in date of manufacture AND recently calibrated to traceable, verifiable standards.
You can, however, determine a fairly repeatable offset between the two with extensive measurements over time between the two, provided all variables per unit remain the same.
If your use is casual and not critical, as long as you use a repeatable user calibration scheme on each densitometry, you should be fine.
BTW, I would purchase a nice new 21 step wedge to plot response curves for each machine and to use as a calibration tool. Base plus fog or just clear leader is NOT a good calibration target; you need a verifiable density you can share between the machines. Besides, you don't subtract out the density of the base as a general rule when measuring density of a film strip; the base IS part of the overall density.
Frank