I have the Ilford 600H system. It is also known as the Durst Multigraph. Do you have the maintenence manual for it?
According to the manual, the Cont Err means:
"An unrealistic contrast value has been set in the grade table for the paper grade required."
The suggested corrective action is:
"Check the paper/grade table and re-enter revised values as necessary or select Contrast and press Clear to reset the factory settings.
Try selecting Contrast and pressing CLEAR to reset the factory default values. If that fails, most likely the battery is dead, and those values may have been lost.
I expect the noise associated with the dichroic filter movement is a result of it trying to move to a value outside its normal range.
I would avoid hitting EXPOSE (and thus avoid the noise) until you get the Cont Err problem fixed. Attempting an exposure could damage the mechanism.
The backup battery should be a 'coin' cell inside the control unit. Hopefully it's in a socket, not soldered directly to the board.
Lithium coin cells have a nominal service life of ten years, but may fail sooner. The unit might be more than ten years old.
I have a 500H, but it's quite different. It has two 300w lamps, each with its own fixed dichroic filter, one green and one blue.
- Leigh