I have now fixed the fault having spoken to
a very helpful lady on the JOBO Analogue Support desk in Germany.
The fault was due to a malfunctioning Flow sensor used to detect fluid transferring through the lift arm.
The fix is as follows
Connect machine to power, but leave switched off.
Press Enter, Start and Power on buttons simultaneously. This brings up
the Service Menu on which there are a great many things - 14 items I
think.
Scroll down to No 11- Fill sensor. Press Enter, then toggle using up or
down arrow to change from With to W/O. Press Enter.
Press Power Off button.
Repair Complete!!!!
I have just run a test with water in all 3 bottles and the
machine pumped all chemicals in all bottles through with no errors.
It does mean that I have to fill with correct volume of chemicals for
each run. However this was part of my routine anyway in order to
conserve E6 chemicals.
For those with ATL2X00 machines, this Service menu could be the place
to start when things go wrong. I shall look there first in preference
to experiencing similar hassle to yesterdays palaver when I stripped the machine down, reassembled it and found that the fault was still present. I only hope that
I haven't written this in haste!
Here is the link to the Google Darkroom users group where I found several very helpful people who contributed to the fault finding.
http://groups.google.co.uk/group/rec.photo.darkroom?hl=en