You might find it worthwhile to join and post your question on the Imacom Users group.
ImaconUsers groups.io Group
A discussion group for users of Imacon and Hasselblad high-end scanners. Topics include using and maintaining the scanners.groups.io
Gerry Yaeger, who moderates the group, also has his own website with a wealth of useful information for Imacon users
You might find it worthwhile to join and post your question on the Imacom Users group.
Welcome to Photrio @anywayform!
I've taken your image and boosted contrast extremely in order to emphasize the defect:
View attachment 384413
The main problem appears to be on the green/magenta channel mostly, showing up as a single, broad band, and it seems to extend all across the frame. The reason why you're mostly seeing it in skies is simply because it's more apparent against the lack of any other features or color separation - it just stands out there, more. It's not isolated to skies, however.
There is some additional vertical striation that may or may not be related.
Thanks for posting that link - very useful indeed! A quick search on their group turns up e.g. this: https://groups.io/g/ImaconUsers/topic/101795211#msg1556
Sadly, no solution offered there.
This one seems also relevant: https://groups.io/g/ImaconUsers/topic/83759364
In that thread, there's quite some discussion about the film itself being the problem (e.g. processing-related), but that's not the case here. The pattern of your color anomaly isn't consistent with common processing defects, and moreover, you've already established that it occurs with other negatives as well. That is to say, you do in fact have a problem with this negative as well, which is in the top right corner:
View attachment 384414 (note somewhat wiggly, lighter vertical stripe about 1/4 from the left edge of this snippet).
That's a film/negative-related issue and I'd ignore it in the context of troubleshooting the scanner problem.
Maybe an initial step in problem solving is to establish the exact nature of the defect; so far in the example image we have:
1: Some dust; not really a concern
2: A processing-related problem manifesting as a lighter wiggly line running vertically through this frame; unrelated to scanning.
3: Some fairly subtle, narrow vertical banding along the entire width of the frame; likely scanner-related
4: A broad magenta/green band running vertically along the frame, which is virtually guaranteed to be scanner-related.
I'd scan some clean, known-good negatives (e.g. lab-processed C41) with large expanses of fairly even tone (skies etc.) so you get a good view on what's happening exactly.
As to potential causes, my first thoughts would go in the direction of something related to the light source for the broad magenta band, and dust on either the light source or CCD assembly for the narrow vertical stripes.
No apology needed, it's exciting to work on such cool scanners, troubleshooting is part of the thing! Here is a link to the Imacon 848 Service Manual that I've been using thus far.No problem; sorry that there's no solution just yet, but at least we can try to work together in locating potential causes.
As to the tubes: if these are the fluorescent tubes that are commonly seen in especially older scanners, I don't really see how a defect in a tube would cause this partial color anomaly. Unfortunately, I'm insufficiently familiar with how this scanner is constructed internally to make any more educated guesses. Is there a page with photos of its internals or perhaps even a service manual out there somewhere that I could have a look at? (recently @gary mulder posted a photo of the lens stage inside a Flextight scanner - I'm not sure what type he uses and if he happens to have other photos of the internal configuration)
There's a couple of pieces of glass inside the machine that can pick up dust and cause banding - they aren't easy to get to - either a fairly full strip down or a slightly hair raising dive in with sensor swabs while running a holderless scan.
Ah, thanks; that helps already. It would be very useful if you could manage to follow the instructions on page 5-1 and post the RGB plot that you obtain this way. The banding problem may show up here. Either way, it might help to track down the issue.
That's a great idea! Unfortunately I'm not seeing the "Monitor" option on my maintenance menu. I'm thinking its a software difference as I'm using a 2011 Macbook Pro to run FlexColor 4.8.13. I'm digging around to see if there's another way to pull it up on my version of the software. I'll update once I've found a solution!
Unfortunately I'm not seeing the "Monitor" option on my maintenance menu.
The service manual method (Pg 5-1) to get the monitor option to show up works fine on a similar vintage iMac.
The bits of glass in question are just below where the neg carrier travels. They can be reached with a sensor swab from the front if you set it running on a blank scan. While this was explained by the person who still maintains these machines, only do this at your own risk!!! Getting them out requires a much fuller strip-down than the linked service manual.
You need to type "debg" in the main program window. The monitor option only appears after you've done this.
The service manual method (Pg 5-1) to get the monitor option to show up works fine on a similar vintage iMac.
The bits of glass in question are just below where the neg carrier travels. They can be reached with a sensor swab from the front if you set it running on a blank scan. While this was explained by the person who still maintains these machines, only do this at your own risk!!! Getting them out requires a much fuller strip-down than the linked service manual.
With the dSLR scanning setup, you apparently have some light bleeding along the edges of the film. That's a different matter; let's set that aside for now.
The Flextight sensor plot doesn't look perfect to me:
View attachment 384620
Note how red and green bump upwards somewhere halfway the plot. This will make it impossible to get the same color calibration across the entire image area. Whether this deviation is big enough to present problems, I can't say. Color negative is a rather low-contrast image, so any slight scanning problem will amplify in post processing when contrast is increased dramatically to fit the normal tonal scale. I can imagine that a small deviation could thus work out as a fairly big/visible problem.
I wonder if you could use the 4x5 holder and mount your strip of medium format film on either side of it, and scan it as a 4x5" frame. This should ensure that all the scanning would happen to the left or the right of that bump in your sensor plot. You could even experiment with shifting the strip of film around to see if the color band somehow aligns with that odd bump in your sensor plot. Note that you may/will have to mask the open part of the 4x5" frame to eliminate stray light problems. For this test, the reduced effective resolution you'll end up with should be no problem.
I am a bit suspicious of the film stock Cinestill 50D
Recalibrating is definitely something worth a shot.
Why would this be? I'm not aware of any notable quality concerns apart from the inherent lack of anti-halation.
Maybe for now you're better off 'scanning' with the digital camera. Looks like you've got that process under control reasonably well and it's certainly a whole lot faster than the Flextight.
Btw, I've borrowed an old Flextight from a friend so I might be able to do some testing later on. I'm currently waiting for a SCSI adapter and cable for my PC to arrive; it's a bit uncertain I'll get the combination to fire up, but let's see. The scanner came with an old but functional WinXP laptop so worst case I can use it with that, but I'd rather just hook it up to my regular desktop.
For 6x7, isn't the Flextight limited to 3200dpi? That's not much more an epson flatbed. A camera can definitely outperform that in terms of detail with the right equipment and technique.
For 6x7, isn't the Flextight limited to 3200dpi? That's not much more an epson flatbed. A camera can definitely outperform that in terms of detail with the right equipment and technique.
With the dSLR scanning setup, you apparently have some light bleeding along the edges of the film. That's a different matter; let's set that aside for now.
The Flextight sensor plot doesn't look perfect to me:
View attachment 384620
Note how red and green bump upwards somewhere halfway the plot. This will make it impossible to get the same color calibration across the entire image area. Whether this deviation is big enough to present problems, I can't say. Color negative is a rather low-contrast image, so any slight scanning problem will amplify in post processing when contrast is increased dramatically to fit the normal tonal scale. I can imagine that a small deviation could thus work out as a fairly big/visible problem.
I wonder if you could use the 4x5 holder and mount your strip of medium format film on either side of it, and scan it as a 4x5" frame. This should ensure that all the scanning would happen to the left or the right of that bump in your sensor plot. You could even experiment with shifting the strip of film around to see if the color band somehow aligns with that odd bump in your sensor plot. Note that you may/will have to mask the open part of the 4x5" frame to eliminate stray light problems. For this test, the reduced effective resolution you'll end up with should be no problem.