Howtek D4000 Drum Scanner Lines

CK341

A
CK341

  • 0
  • 0
  • 25
Plum, Sun, Shade.jpeg

A
Plum, Sun, Shade.jpeg

  • sly
  • May 8, 2025
  • 0
  • 0
  • 32
Windfall 1.jpeg

A
Windfall 1.jpeg

  • sly
  • May 8, 2025
  • 2
  • 0
  • 31
Windfall 2.jpeg

A
Windfall 2.jpeg

  • sly
  • May 8, 2025
  • 1
  • 0
  • 30
Marsh, Oak Leaves.jpeg

A
Marsh, Oak Leaves.jpeg

  • sly
  • May 8, 2025
  • 0
  • 0
  • 29

Recent Classifieds

Forum statistics

Threads
197,610
Messages
2,761,920
Members
99,416
Latest member
TomYC
Recent bookmarks
0

Fixcinater

Member
Joined
Dec 11, 2008
Messages
2,500
Location
San Diego, CA
Format
Medium Format
I've been slowly working my way through reviving a Howtek D4000 with the COSMYK 860 chip. Mirrored Drive Doors Mac tower running OS X 10.4.11 and Silverfast in demo mode. Firmware has been updated by me to R535.

Replaced:
Bulb, twice over
Drive belt
O-ring on the drum driveshaft/mounting point
Drum pads
Added transformer based power supply filtering between wall AC and scanner, no change
Cleaned calibration area on drum (was clean but just making sure no tape or residue existed)
Lubricated the lead screw (was bone dry when it came to me)
Cleaned the lead screw thinking I had over-lubricated it, no change
Made sure temp in room is higher than 50F, no change
Kept scanner on for multiple hours, no change

Attached are two files that show the difference between 1K dpi and 2K dpi, otherwise all settings exactly the same.

Still getting this awful line whenever I scan something at 2K dpi or higher. I've attached two small crops showing the lines.


Potential issues I still see:
-Lamp intensity fluctuates during scan. Does not seem to have a pattern, whereas the lines in the scans do seem to be regular.
-When I lubed the lead screw, the variance between the darkest parts of the lines and lightest seemed to diminish.
 

Attachments

  • howtekD4K@1Kdpi.jpeg
    howtekD4K@1Kdpi.jpeg
    47.5 KB · Views: 325
  • howtekD4K@2K_DPI.jpeg
    howtekD4K@2K_DPI.jpeg
    51.2 KB · Views: 321
OP
OP
Fixcinater

Fixcinater

Member
Joined
Dec 11, 2008
Messages
2,500
Location
San Diego, CA
Format
Medium Format
Did some more work on it between last posting and now:

-Cleaned and cleaned again the pulleys for the drive belt.
-Cleaned and lubed the lead screw again.
-Cleaned and lubed the tailstock bearing.
-Cleaned the same drum I had been using.
-Re-did the DIY pads I had been using, slightly thinner so they were all within the small recess as the factory pads are. Before, some were slightly too wide which led to the drum not centering correctly on the tailstock.
-Cleaned the dust out from behind the FOPI, on that lens.
-Cleaned the lens on the other side of the FOPI, illuminator lens?

Now does 2K dpi scans with no hint of lines so that's a big relief.
Next thing to suss out is now when I set 4K DPI in Silverfast, it always returns an F702 error when checking lumens. At 1K, 2K there is never a problem but at 4K each and every time it pulls the error code.

Sample shot attached:

Wartime black paint Anniversary Speed GraphicBlack paint Kodak Ektar 127/4.7HP5 in HC110 1:100 semi-stand
Blue painter's tape, dry mounted
Scanned at 2K DPI

Here is the result:

Howtek D4000 Drum Scanner Lives Again by Dead Link Removed, on Flickr
 

Pasto

Subscriber
Joined
Dec 15, 2004
Messages
864
Location
Montreal
Format
Multi Format
Hi, just came upon your thread. Am wondering how things turned out with your D4000? I've been going through a similar revival program for about a month now. Your experiences may be helpful. Just for background:

Scanner: Howtek D4000
Computer: AMD Phenom 9650 quad core, Windows XP professional SP3, Adaptec 2906 SCSI adapter
Software: Trident, Aurora, and Silverfast 6.6 all in demo mode.

Problem 1: Was getting multiple lamp/mirror R (or RGB), Error F719.
Solutions (?): Disassembled all control boards from scanner, cleaned all accessible connectors including slots for control boards, and grounds with connector cleaner, fixed problem with my drums' end-caps becoming unglued from cylinder causing index mark on left end-cap being out of alignment with top scribe line in calibration area, replaced bulbs, kept scanner on for several hours. This problem is largely solved now, I hope.

Problem 2: Since fixing 1, I've been able to get good scans (both reflective and transmission) with all the scanner software. Then all of a sudden received a calibration...F702 error, and continue to do so.
Solution (in progress): change drive belt, as well as replace the drive motor and encoder connectors (as per Evan's instructions at Aztek). Now waiting for parts.
 

Tropper

Member
Joined
Jan 22, 2019
Messages
9
Location
GA
Format
35mm RF
Sounds like a very delicate process to restore that stuff. Too bad the pics aren't working anymore.
 

Pasto

Subscriber
Joined
Dec 15, 2004
Messages
864
Location
Montreal
Format
Multi Format
Actually it's not so bad. Sorry that the original poster is no longer posting his experiences. Since changing the connectors for J12 and J1 on topmost board the scanner runs like new. I'm now sourcing the fluid and tape required to try wet mounted scans. I've been running dry scans in silverfast demo.
 
OP
OP
Fixcinater

Fixcinater

Member
Joined
Dec 11, 2008
Messages
2,500
Location
San Diego, CA
Format
Medium Format
Just saw that there were some responses here. Glad to see yours is now up and running. I've been unable to make any progress on mine after it started throwing another error specific to a board, and life has gotten too busy to sink in component level diagnosing for me. I'm now back in southern California so I may be dropping it off with Aztek to get it running.
 

Scott J.

Member
Joined
Apr 18, 2017
Messages
150
Location
Wyoming
Format
Large Format
I wanted to reply with the results of troubleshooting a recent and virtually identical problem with my Howtek SM4500. A little background...

I'm the third owner of this scanner and have had it for about 1.5 years. When I initially bought it, it had a persistent problem with an F719 G error that prevented it from performing any kind of scan (which the previous owner was completely upfront and honest about). After a few weeks of familiarizing myself with the scanner and corresponding with Evan at Aztek, I eventually traced the problem to a bad high voltage power supply (HVPS) on the green channel. To diagnose this, I swapped the suspect HVPS with each of the other channels (R and B) and then tried scanning. When connected to the other color channels, the F719 error would get reported for which every color channel the suspect HVPS was connected to. Evan agreed that it was likely that the HVPS on the green channel was the problem, but cautioned that it might still be something else in the signal chain. I took a slight gamble and purchased a replacement HVPS from Aztek (around $560 in 2018, as I recall), installed it (pretty simple), and was able to start scanning. I run the scanner with DPL 8.1 on a 64-bit Windows 10 machine and have been quite happy with the results (mostly 4x5, but the occasional 6x17, 6x9, and 6x7). The only recurring and slightly annoying problem I've had is that I usually get a few F702 errors in a row when attempting preview scans after initial power up of the scanner (I only scan a couple times a week when not travelling for work). After 3-5 failed attempts at previews, the scanner will eventually successfully perform preview scans, detail scans, aperture checks, and final scans without any trouble. It's almost as if it just needs a little coaxing or warming up to get it past the F702 errors, but when it does, it runs just fine for the rest of the scanning session (which can last several hours). So, that's a nuisance, but hasn't been fatal (knock on wood).

Fast forward to a few weeks ago. My transparency lamp went bad so I ordered a replacement (got an OEM replacement on Amazon for about $16). The scanner started scanning again successfully. Then, after a few days, I started noticing horizontal lines in my 4,000-dpi scans. In other words, lines that are parallel to the direction of drum rotation. Below is an aperture check scan at 4,000 dpi and 19 microns from a small area of a piece of 120 Fuji Pro 400H. (It's not a great sample, but you get the idea.)

APERTURE_1.jpg


The horizontal lines appear one-pixel wide. When I scanned at 2,000 dpi or lower, the lines were not apparent. It seemed to me that the lines indicated either a variation in lamp output, an aberration in the white/black-point calibration (which I understand the scanner does on every rotation of the drum), or a kind of inconsistency to the drum rotation that results in a kind of "pixel offset" error on each rotation. These lines look similar to what Fixcinator reported in his original post. My guess is that the reason the lines didn't show up at 2,000 dpi is because at lower resolutions the pixels either become wide enough that they "conceal" the lines or that the higher drum speed at lower resolutions keeps the drum from "wobbling" as much (or some combination thereof). In any case, I tried the usual stuff: swapped lamps, checked the focus, checked the optical alignment, greased the lead screw, changed drum pads, tried a different UPS, checked the drum for scratches and debris in the calibration area, tried a different piece of film... and to no avail. I did a little reading in the manual and on the Scan Hi End group on Yahoo about what might cause these kinds of lines, and it seemed that the consensus pointed to something having to do with the stability of the drum rotation or the way the scanner reported the drum's position to the scanning software (which, if in error, could cause this sort of line-by-line weirdness).

Long story short: I got the scanner working properly again (no more horizontal lines). The last two things I did were done simultaneously, so I can't say for sure which did the trick, but they were: 1) gently brushed and blew off the carriage code wheel and detector (Pages 5-22 and 5-23 in the SM4500 service manual); and 2) put on a new drive belt. Everything is back to normal now. If I had to guess, the new drive belt is what did the trick as the carriage code wheel looked pretty clean to me (be extremely gentle if/when cleaning it).

My hypothesis is that, because my scanner occasionally sits unused for extended periods of time (sometimes 4-6 weeks), the belt begins to develop an uneven strain along one axis because of the way it's sitting in the scanner on the two pulleys. If you're scanning frequently (e.g., a couple times a week), this is probably enough to ensure that the belt ends up resting in a different position in the two pulleys every few days, such that the strain is more evenly distributed around the belt's circumference. In my case, the uneven strain concentrated along one axis was probably causing the belt to slip a little bit on each rotation (or maybe every few rotations), thereby resulting in the weird, pixel-wide offset. The amount of slippage was presumably not severe enough to show up at the lower resolutions, or maybe the higher rotation speed meant that the drive belt wasn't slipping as bad. Looking at the bad belt and comparing with a new one, I can't see anything obviously wrong with it other than the fact that it's average radius is slightly larger than that of a new belt, but I'd be willing to guess that there is some degree of disproportionate axial strain aligned in one direction. I may adopt the practice of taking the belt out of the scanner if I know it's going to sit for long periods of time.

Hope that helps.
 

brbo

Member
Joined
Dec 28, 2011
Messages
2,025
Location
EU
Format
Multi Format
Nice to hear you scanner is still/again feeling well. I always remove the belt after scanning and store it in a ziploc bag.
 