Hi Dave, judging from your response, if both meters read the same with the dome to the side, then the likely suspect is the dome. Perhaps a small insect managed to get inside the dome and create an obstruction.
If you wish to inspect, you can remove the single screw on the back plate and remove the plate. Take a small piece of scotch tape and tape the meter zero plastic slotted screw in place so it doesn't fall out. Once inside you will find four brass slotted screws at the corners of the PC board. Loosen these screws a couple of turns or so and you can possibly create enough slack to remove the dome for inspection/cleaning.
The meter's clamshell is held together by these four screws AND three wires coming from the nulling pot/dial and to completely separate the two halves you must un-solder these wires. If you decide to try and remove the dome and not un-solder the three wires, do not exert much force in trying to create enough slack to separate the two halves enough to extract the dome as the wires to the nulling pot will break. Usually you can can gently create enough of a gap at the dome to remove it. The dome is flat on the 'bottom' side and has a 'spring loaded' plastic bump on the top side. Push the dome down into the slot in the bottom half and tip the dome out for cleaning. Once cleaned, re-insert the dome by re-inserting the flat side in the bottom slot, slightly compress the spring loaded bump and tip back into its slot. Re-tighten the four brass screws gently, remove the scotch tape and re-install the back plate with its screw. When re-installing the back plate, don't let the slotted plastic meter zeroing screw fall out of its hole. It's tricky to get it back into the slot on the meter movement's adjustment arm. That's it!