The only four patches to measure ( RGB ) for plotting onto a standard Y55 form are ....
D-min
LD ( 3 ) low density 'speed'
HD ( 5 ) high density 'colour'
D-max
Y-55 form is downloadable here .... https://imaging.kodakalaris.com/sites/default/files/wysiwyg/retailers/chemistry/techpub/y55.pdf
The other patches are designed for full sensitometric plotting which are not generally required.
John S
That's right, a processed reference strip ( as pictured above ) is supplied for you to measure using a calibrated densitometer.
After applying the correction factors included, these become your aim points.
John S
Don't forget you must use the 'status A' filters on your densitometer, and if you're just starting to do this, you should also include 'control limits' and 'action limits' on your control form.