On a scan of a negative or a positive?- It's a lighter line and not a darker line, so I'm unsure that it's dust.
how to get into the actual sensor?
On a scan of a negative or a positive?
You don't want to and it's also virtually certain that you don't need to. In 99.9% of the cases, the kind of problem you're describing is down to fouling of the platen glass or the glass in the transparency lid. Note that when you're working with a flatbed scanner for transparencies, there are 4 glass surfaces that at some point need cleaning. This means in practice (partly) disassembling the unit and cleaning the glass. I've not worked on the 10000XL, but on the smaller Epson flatbeds it's a fairly simple matter of removing some screws and opening the housing (of either the box itself or the transparency lid) and then lifting out the glass. Be careful not to scratch it.
Ok, thanks for clarifying. You cleaned the glass top and bottom?
PS: by cleaning, I mean properly cleaning. Not vacuuming the unit out. This will likely not be sufficient.
Thanks, I see what you mean. It's subtle, but it does appear to be colored and not monochrome and it appears to be fairly sharply delineated as well. Could be dust on the sensor alright as you initially expected, but I'm not 100% certain.
This suggests some dirt/smudge on the calibration strip. It's a strip tat is read by the sensor at the beginning of each scan, to calibrate out the deviations from uniform response of the pixels. Don't know about the XL10000, but on my V700, that strip is on the underside/top of the top plate, just above the scanning glass. I think (IIRC) it is covered by the glass. Still on my V700, removing the top plate is easy, but separating the glass from its frame to access the calibration strip is not trivial. Hopefully you have a smudge on the bottom (top once you have removed and flipped it) of the glass, above the cal strip.Any idea why it would be lighter as opposed to darker?
This suggests some dirt/smudge on the calibration strip. It's a strip tat is read by the sensor at the beginning of each scan, to calibrate out the deviations from uniform response of the pixels. Don't know about the XL10000, but on my V700, that strip is on the underside/top of the top plate, just above the scanning glass. I think (IIRC) it is covered by the glass. Still on my V700, removing the top plate is easy, but separating the glass from its frame to access the calibration strip is not trivial. Hopefully you have a smudge on the bottom (top once you have removed and flipped it) of the glass, above the cal strip.
If you find there is dirt between the glass and the strip, I would suggest to overlay the strip (and glass) with a strip of smooth white paper. Something like satin or semi-matte inkjet paper. Be aware that this strip defines the white point; so strange things might happen if your inkjet paper has optical brighteners...
Good luck