Folks,
Apologies if this has already been discussed before - if it has been, I overlooked the thread.
I am trying to calibrate my new RH Analyser Pro (Fantastic machine, btw!) and am stuck at the very first step. For reproducibility I "reset all variables": new light bulb (of which I have a second one of the same make & model), new developer, etc. I also use the lens and condenser I use for all my printing i.e. Rodagon 90 and Bimacon 75. I use the same for both 35mm and 6x6. Enlarger is an M805. Light bulb is a "Dr Fischer 150W Opallampe" available at Fotoimpex, 150W being the default for the M805.
Even uncalibrated, the Analyser has provided good and consistent results with the setup described above, so it's just a matter of fine-tuning.
Measurements before printing are done without any MultiGrade filter, then a filter is installed after the indication of the Analyser, as indicated in the user's manual (unless I misunderstood it!).
TrueQuestion 1. Is my understanding correct that:
a. Calibration is done without any negative in the neg holder
True alsob. Before doing the speed test strips, a measurement is done with the Analyser. This measurement is done without any MG filter (as for normal prints). The measurement should bring an exposure time between 10 and 20 seconds.
c. Then, a series of test strips based on the exposure determined in (b) are done with each grade filter.
True(d. Side note, only grades 00, 0, 1, 2, 3, 4, 5 are measured, although the Analyser supports intermediate grade values as well)
You need to slide a piece of ND filter in the filter slot to bring the light level down so that the length of test is in the ballpark.Question 2. Even with the head in its highest position and the lens at f/22, the measured exposure is 2.88 seconds - way too short to make a test strip. (the Analyser wouldn't let me do it even if I wanted to).
What shall I do? Not sure an ND filter is a good thing here... shall I make a "neutral grey" negative and use that for calibration? Or shall I dump the 150W bulb and use 75W instead? (but the 150W has been working great as soon as a negative is in place). Any advice?
Thanks!
ETN
Well...I typed that from memory and you are correct. Whatever you do to reduce the light doesn't factor in. The directions also said you can use a piece of unexposed but processed film leader or ND filter or even dialing in equal amounts of Y-C-B if you are using a color enlarger to reduce the light level to be correct in length.Hello Rpavich,
Thank you. Very helpful. The last thing I don't understand is the necessity of taking the ND filter into account into the offset calculation. Differences in luminosity due to lens f/stop or head position (and yes, I cranked it all the way up, btw) are not taken into consideration either. Giving second thoughts to the entire process since yesterday, what we are calibrating here is the luminosity value measured by the Analyser, vs certain tones on the paper. So everything on the enlarger side is theoretically out of the equation. But other variables such as paper, developer etc are not. Does this make sense?
If this is true, we should not take the ND filter into consideration, as long as it is in place BEFORE doing the initial measurement, what do you think?
As a result, the ND filter does not need to be precise to a known density. I will try to do my own filter by photographying an evenly-lit white wall at different exposures levels. I'll use that for a first calibration round and analyze the results. Will report my findings here.
Thanks & a good new year 2017!