In fact, IIRC, that's one use of this meter that Darkroom Automation (AKA Nicholas Lindan) recommends. You can put the meter under an enlarger, and zero the meter by pressing its delta-mode button. Then, when you put a negative over the meter's sensor, it tells you its density. Viola! A transmission densitometer.Seems with a simple downward-pointing light, you could make a transmission densitometer
But I struggle with reading black values. I can't get higher than 5.8 stop for maximum paper black (relative to paper white).
A reflection densitometer needs two calibration points:
- White point calibration - in this case paper white works to zero 'delta' mode. The readings will be in density over the paper base;
- Black point calibration - ideally made with the densitometer pointing into the infinite void with the light source on - this removes the effect of internal reflections in the densitometer.
The Darkroom Automation enlarging meter was never designed for conversion to a reflection densitometer, so there is no way to take a black point calibration reading.
It would be interesting to know the reading you get when taking a reading pointing into the darkness - the darkroom should be dark enough.
I was inspired by this and because I'm not very good with wood, I printed a copy on a 3D printer:
View attachment 392447
But I struggle with reading black values. I can't get higher than 5.8 stop for maximum paper black (relative to paper white).
The amount of light does not affect this, specular reflection has a place to escape.
What should I focus on?
I like your design. Good work.
5.8 stops is a density of 1.75. Are you sure your paper is darker than that? Ilford paper reaches 2.1, but my Foma paper reaches only 1.8 or 1.9.
Is your paper glossy? A semi-gloss or matte surface might reach only 1.75.
Here is a simple test: Hold the paper at a slight angle on your device, causing the specular reflection to be closer to horizontal. If that improves black measurement, then a little of the specular reflection is getting onto the sensor when the paper is horizontal.
Mark
5.8 stop is about all you can expect in reflection mode.
For comparison a Stouffer scale for reflection has 13 numbered half stop steps and you can’t always measure the last one.
I use Fomabrom Variant 111. Glossy, but not too much.
That should hit 2.1 logD alright.