Here's an accurate reflection densitometer that costs $120, which is far less than anything else on the market. It consists of an enlarger meter from Darkroom Automation, and a home-made attachment containing an LED.

1-OnMeter.jpg


It's accurate! Measurements from this simple attachment differ from my calibrated Stouffer reflection step-tablet by at most 0.02. First, here's the enlarger meter by itself. This picture was taken from http://www.darkroomautomation.com/em.htm .

2-Meter.jpg


I taped a white piece of paper to the top of my meter to give me a larger view of the surroundings, making it easier to find a spot I want to meter on the easel. This addition is unrelated to the densitometer.

You will need a high-power LED, which you can take out of an LED torch (flashlight). This is a good use of a torch that was ruined by leaking batteries. Below, you see that I desoldered and removed one LED.

3-TorchParts.jpg


The torch had a clear plastic cover, shown on the right side of the photo above. Keep it. You will use it to hold the print flat on top of the densitometer. The light spot from the LED is so bright that the image shows through the back of the print, so using this clear cover allows you to see exactly what spot on the print you are measuring.

The sheet metal merely holds the wood block in the correct position. You can use pieces of wood instead if you don't have access to sheet metal. In fact, I recommend building a rectangle that barely touches the meter on all four sides, as that will hold the LED-block in place in all directions. My poorer design allows the attachment to slip northward of the meter.

The wood block has a corner-to-top 45-degree hole through which the LED shines up to the top, and a vertical hole over the meter's sensor. The photo below looks directly down the 45-degree hole. I suggest making the vertical hole 6mm in diameter (1/4 inch), and the 45-degree hole just large enough to allow the upper portion of the LED to slide into it. The bottom of the LED has a flared lip that can remain outside the hole. The reason for making this hole as small as possible is to minimize the spot-size at the top, thus minimizing the diameter of the troublesome specular reflection described below.

4-45DegView.jpg


The diagram below shows a side-view of the light-path through the wood block.

5-Diagram.jpg


The LED at the lower left shines its beam up the 45-degree hole, which reflects off the print on the top. This reflection consists of a diffuse component which travels down the vertical hole and onto the sensor, and a 45-degree specular component which can only cause flare. I cut a notch in the side of the block to allow the specular beam to escape without causing flare. I painted the area around the top black to cut flare even more. Also, the first photo above shows a vertical notch cut in the top to allow the specular reflection to escape without touching any wood, causing flare. That's the key: remove all wood from the path of that specular reflection.

Flare is the enemy when measuring dark areas, because a little flare will cause near-blacks to appear significantly lighter. Those two notches conquered flare: My measurement of the darkest patch on the calibrated Stouffer tablet was 0.01 higher (darker) than Stouffer's measurement. Flare would have made my density measurement lower, not higher.

To operate the LED, find a 5-volt switching power-supply, such as those for consumer electronics or USB. Obtain several 100-ohm resistors rated for 1/2 watt or higher. These can be purchased cheaply from electronics suppliers such as digikey.com, mouser.com, and newark.com. I found that 300 ohms (three resistors) in series with the LED is perfect: the spot's brightness reflecting off white paper is near the meter's upper limit. You might find that 200 or 400 ohms is best with your LED.

I should mention that the meter measures in stops, which you must multiply by 0.301 to get log D. Instead of doing that, I convert my desired densities to stops by dividing them by 0.301.

The densitometer is stable. When measuring densities, I always zero the meter on white paper, putting the meter in "delta mode". After 10-15 minutes of measuring many test-patches for H-D curves, I always re-check the zeroing by measuring white again. It always remains at 0.00. That's stability. And a max 0.02 error is good accuracy.

Mark Overton
 

Bill Burk

Subscriber
Joined
Feb 9, 2010
Messages
9,290
Format
4x5 Format
Seems with a simple downward-pointing light, you could make a transmission densitometer
 
OP
OP

albada

Subscriber
Joined
Apr 10, 2008
Messages
2,172
Location
Escondido, C
Format
35mm RF
Seems with a simple downward-pointing light, you could make a transmission densitometer
In fact, IIRC, that's one use of this meter that Darkroom Automation (AKA Nicholas Lindan) recommends. You can put the meter under an enlarger, and zero the meter by pressing its delta-mode button. Then, when you put a negative over the meter's sensor, it tells you its density. Viola! A transmission densitometer.

If you put the negative in the enlarger's carrier, and with the meter on the easel, you will measure the light that would strike the paper, which is determined by negative-density plus flare. By putting the neg on the meter, flare is eliminated, yielding an accurate density-measurement.

Mark Overton
 
OP
OP

albada

Subscriber
Joined
Apr 10, 2008
Messages
2,172
Location
Escondido, C
Format
35mm RF
Here are some H-D curves for Ilford MGRC Deluxe (aka "RC V") paper obtained by contact printing the Stouffer T3110 step wedge, and measuring densities using this semi home-made densitometer.

Curves13a.jpg


The numbers on the X-axis correspond to Stouffer steps, spaced 0.10 logD apart. The Y-axis is density in stops relative to paper-white (I zeroed the meter on white paper). Row 1 is grade 5, row 5 is grade 2, row 8 is grade 0. The curves were supposed to cross step 8 at the same point (which is zone 5, mid-gray), but they're a little off because I created the zone alignment table by matching tones by eye, before I made this densitometer. This densitometer is wonderful.

Mark Overton
 

blacksquare

Member
Joined
Sep 8, 2015
Messages
74
Location
Czech Republic
Format
Large Format
I was inspired by this and because I'm not very good with wood, I printed a copy on a 3D printer:

IMG-20250105-WA0002.jpg

But I struggle with reading black values. I can't get higher than 5.8 stop for maximum paper black (relative to paper white).
The amount of light does not affect this, specular reflection has a place to escape.

What should I focus on?
 

Nicholas Lindan

Advertiser
Advertiser
Joined
Sep 2, 2006
Messages
4,243
Location
Cleveland, Ohio
Format
Multi Format
But I struggle with reading black values. I can't get higher than 5.8 stop for maximum paper black (relative to paper white).

A reflection densitometer needs two calibration points:

  • White point calibration - in this case paper white works to zero 'delta' mode. The readings will be in density over the paper base;
  • Black point calibration - ideally made with the densitometer pointing into the infinite void with the light source on - this removes the effect of internal reflections in the densitometer.

The Darkroom Automation enlarging meter was never designed for conversion to a reflection densitometer, so there is no way to take a black point calibration reading.

It would be interesting to know the reading you get when taking a reading pointing into the darkness - the darkroom should be dark enough.
 

blacksquare

Member
Joined
Sep 8, 2015
Messages
74
Location
Czech Republic
Format
Large Format
A reflection densitometer needs two calibration points:

  • White point calibration - in this case paper white works to zero 'delta' mode. The readings will be in density over the paper base;
  • Black point calibration - ideally made with the densitometer pointing into the infinite void with the light source on - this removes the effect of internal reflections in the densitometer.

The Darkroom Automation enlarging meter was never designed for conversion to a reflection densitometer, so there is no way to take a black point calibration reading.

It would be interesting to know the reading you get when taking a reading pointing into the darkness - the darkroom should be dark enough.

I wouldn't have even thought about this possibility if I hadn't come across this topic. And if it works for someone, why not give it a try.
And at low densities it worked well (enlarged Stouffer wedge).

Without the laid paper on top, it's just "Und"....the light from the LED has nothing to reflect on the sensor.
 
OP
OP

albada

Subscriber
Joined
Apr 10, 2008
Messages
2,172
Location
Escondido, C
Format
35mm RF
I like your design. Good work.

5.8 stops is a density of 1.75. Are you sure your paper is darker than that? Ilford paper reaches 2.1, but my Foma paper reaches only 1.8 or 1.9.
Is your paper glossy? A semi-gloss or matte surface might reach only 1.75.

Here is a simple test: Hold the paper at a slight angle on your device, causing the specular reflection to be closer to horizontal. If that improves black measurement, then a little of the specular reflection is getting onto the sensor when the paper is horizontal.

Mark
 

Bill Burk

Subscriber
Joined
Feb 9, 2010
Messages
9,290
Format
4x5 Format
I was inspired by this and because I'm not very good with wood, I printed a copy on a 3D printer:

View attachment 392447

But I struggle with reading black values. I can't get higher than 5.8 stop for maximum paper black (relative to paper white).
The amount of light does not affect this, specular reflection has a place to escape.

What should I focus on?

5.8 stop is about all you can expect in reflection mode.

For comparison a Stouffer scale for reflection has 13 numbered half stop steps and you can’t always measure the last one.
 

Attachments

  • IMG_2133.jpeg
    IMG_2133.jpeg
    386.2 KB · Views: 33

blacksquare

Member
Joined
Sep 8, 2015
Messages
74
Location
Czech Republic
Format
Large Format
I like your design. Good work.

5.8 stops is a density of 1.75. Are you sure your paper is darker than that? Ilford paper reaches 2.1, but my Foma paper reaches only 1.8 or 1.9.
Is your paper glossy? A semi-gloss or matte surface might reach only 1.75.

Here is a simple test: Hold the paper at a slight angle on your device, causing the specular reflection to be closer to horizontal. If that improves black measurement, then a little of the specular reflection is getting onto the sensor when the paper is horizontal.

Mark

Thanks....and I thank you for your idea.

I use Fomabrom Variant 111. Glossy, but not too much. It didn't occur to me that I was trying to bring values closer to Ilford papers

So I tried now adjust the angle a bit, zero the meter on white paper and yes, it improves black measurement a little, about 1.81. So I'll slightly modify the design and won't deal with it any more, for me these edge values are not that important when printing.

Thanks!

5.8 stop is about all you can expect in reflection mode.

For comparison a Stouffer scale for reflection has 13 numbered half stop steps and you can’t always measure the last one.

Well, I'm smarter now. The shades on this R1215 look very similar to a 13 steps portion of my enlarged T2115, which I have in front of me.
I was looking for a problem where there was none.
 

blacksquare

Member
Joined
Sep 8, 2015
Messages
74
Location
Czech Republic
Format
Large Format
That should hit 2.1 logD alright.

Ugh, that's a lot, I'm not even close to that now. I looked in the datasheet and foma says dmax 2.0.
Now if it makes sense to solve it. I want to build a paper sensitivity table/curve for use with a DA Meter and light version of Zone System.
 